[Dsg-rich] RICH backup N2 manifold installation
Tyler Lemon
tlemon at jlab.org
Fri Feb 1 10:11:22 EST 2019
HI Denny,
Just to review what we talked about in the hall this morning and get some clarification from you:
You all are moving the two 12-bottle N2 banks to the top of the pi tower and will run a gas line from there to the RICH N2 panel. Bank #1 will have a regulator set to ~20 psi and Bank #2' regulator will be set to ~10 psi. This will allow gas to be supplied from Bank #1 until its pressure drops below 10 psi, at which point, Bank #2 will start to supply N2.
Using the attached drawing as a reference, you're going to install the new backup manifold at MV4 rather than what is shown in the drawing. The house N2 supply will remain connected to and flowing through MV1.
Once the backup manifold is installed, DSG will run the cabling for the pressure transducers on the manifold to the hardware interlock cRIO and move the house nitrogen from MV1 to the new backup manifold at the connection point labeled "N2 Gas From 96B".
We want to make sure that whatever the plan is for the installation from the manifold, that N2 flow to RICH not be stopped for more than a few minutes. Since we discussed that MV4 will remain valved off until DSG runs the PT cables, can you all be sure to leave the house N2 supply line attached to MV1 and flowing after you all install the backup manifold?
Thanks,
Tyler
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-rich/attachments/20190201/03b9c02a/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: RICH-N2-valve panel-REV3-08Jan2019.pdf
Type: application/pdf
Size: 25424 bytes
Desc: RICH-N2-valve panel-REV3-08Jan2019.pdf
URL: <https://mailman.jlab.org/pipermail/dsg-rich/attachments/20190201/03b9c02a/attachment-0001.pdf>
More information about the Dsg-rich
mailing list