[Dsg-rich] New alarms for RICH N2 main and backup supply pressure

Tyler Lemon tlemon at jlab.org
Wed Jun 12 15:47:55 EDT 2019


Hello,

After last week's issue of losing N2 flow to RICH, we have implemented two new alarms for the status of the RICH N2 supply.

The first alarm is on the main N2 supply pressure. It is set to alert Hall B Engineering if the main N2 supply pressure at the RICH panel drops below 20 psi for more than 60 seconds. This will indicate that the main N2 supply has been lost and the backup system has started to provide nitrogen to RICH. Using data from last week's nitrogen stoppage (attached) as an example, this alarm would have tripped at one minute after the main flow supply valve was closed.

The second alarm is on the backup N2 supply pressure. It is set to alert Hall B Engineering if the backup N2 supply pressure drops below 10 psi for 60 seconds. This will indicate that the backup N2 supply has started to run out. This alarm would have tripped last week as the backup supply started to run out.

Looking at data for the backup N2 supply pressure, we will also look into adding another alarm at ~15 psi. This alarm would indicate that the backup supply has switched to the second nitrogen bank of the backup system.

With these new alarms, we will be notified of any significant change in RICH N2 supply and be able to take action before N2 flow drops to 0 slm.

Best regards,
Tyler
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-rich/attachments/20190612/b6b810b2/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: RICH N2 Flow and panel pressure.JPG
Type: image/jpeg
Size: 197958 bytes
Desc: RICH N2 Flow and panel pressure.JPG
URL: <https://mailman.jlab.org/pipermail/dsg-rich/attachments/20190612/b6b810b2/attachment-0001.jpe>


More information about the Dsg-rich mailing list