[Dsg-rich] [EXTERNAL] RICH High Voltage
Tyler Lemon
tlemon at jlab.org
Wed Jan 22 13:02:28 EST 2020
Hi Marco,
Have the issues you noted in the entry occurred before or is this the first time we've seen it?
Based on the HV mapping at the link below, a lot of the channels with issues are on the same HV card:
Situation 1.) All channels listed are on HV card in slot 13.
Situation 2.) Tiles 46, 47, 51 - 53, and 57 - 58 correspond to channels 25 - 31 of the HV card in slot 12.
https://clasweb.jlab.org/wiki/images/c/c2/Rich_HV_Naming.txt
Could it be that there is a hardware issue with these cards?
Also, from the list at the link, all the affected channels are "next to" each other in the crate so it could also be, like you said, an issue with the EPICS communication where the channel info wasn't scanned or read out correctly due to a comms loss or network error.
Best regards,
Tyler
________________________________
From: Dsg-rich <dsg-rich-bounces at jlab.org> on behalf of Marco Mirazita <Marco.Mirazita at lnf.infn.it>
Sent: Wednesday, January 22, 2020 12:23 PM
To: Valery Kubarovsky <vpk at jlab.org>; marco contalbrigo <mcontalb at fe.infn.it>; dsg-rich at jlab.org <dsg-rich at jlab.org>
Cc: Aram Movsisyan <maram at mail.desy.de>
Subject: [Dsg-rich] [EXTERNAL] RICH High Voltage
Dear all,
I spoke with Aram about a potential issue on the RICH HV that he noticed
while on shift these last days.
I had a look at epics data and made this elog entry:
https://logbooks.jlab.org/entry/3768485
We should discuss if this is a real issue or not.
Best regards,
Marco
_______________________________________________
Dsg-rich mailing list
Dsg-rich at jlab.org
https://mailman.jlab.org/mailman/listinfo/dsg-rich
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20200122/97b3e469/attachment-0002.html>
More information about the Dsg-rich
mailing list