[dsg-hallb_rich] [New Logentry] RICH Electronic Panel cRIO restored

Bob Miller rmiller at jlab.org
Tue Dec 14 10:34:37 EST 2021


I am going to the Hall this afternoon to cool the target.  I will look for a sensor then.  If I find one, I will contact Valery so I can turn the compressor off to change the sensor.

Regards,
Bob

________________________________
From: Tyler Lemon <tlemon at jlab.org>
Sent: Tuesday, December 14, 2021 9:12 AM
To: Valery Kubarovsky <vpk at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>
Cc: Bob Miller <rmiller at jlab.org>
Subject: Re: [dsg-hallb_rich] [New Logentry] RICH Electronic Panel cRIO restored

I do not have any.

I’m not sure if Bob or anyone else from Hall B engineering has done anything yet though.
________________________________
From: Valery Kubarovsky <vpk at jlab.org>
Sent: Tuesday, December 14, 2021 9:10:30 AM
To: Tyler Lemon <tlemon at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>
Cc: Bob Miller <rmiller at jlab.org>
Subject: Re: [dsg-hallb_rich] [New Logentry] RICH Electronic Panel cRIO restored

Tyler,
Any news about the air pressure sensor?
Thanks,
Valery


________________________________
From: Tyler Lemon <tlemon at jlab.org>
Sent: Tuesday, December 14, 2021 09:07
To: Valery Kubarovsky <vpk at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>
Cc: Nathan Baltzell <baltzell at jlab.org>
Subject: Re: [dsg-hallb_rich] [New Logentry] RICH Electronic Panel cRIO restored

Yes, it should be fine to reenable the alarms.
________________________________
From: Valery Kubarovsky <vpk at jlab.org>
Sent: Tuesday, December 14, 2021 9:06:24 AM
To: Tyler Lemon <tlemon at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>
Cc: Nathan Baltzell <baltzell at jlab.org>
Subject: Re: [dsg-hallb_rich] [New Logentry] RICH Electronic Panel cRIO restored

Thanks Tyler!
May we restore the RICH epics alarm control?
Valery


________________________________
From: dsg-hallb_rich <dsg-hallb_rich-bounces at jlab.org> on behalf of tlemon at jlab.org <tlemon at jlab.org>
Sent: Tuesday, December 14, 2021 08:55
To: dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>
Subject: [dsg-hallb_rich] [New Logentry] RICH Electronic Panel cRIO restored

RICH Electronic Panel cRIO restored<https://logbooks.jlab.org/entry/3962660>

  *   Edit<https://logbooks.jlab.org/entry/3962660/edit?destination=email/send>
  *   Delete<https://logbooks.jlab.org/entry/3962660/delete?destination=email/send>

Lognumber 3962660<https://logbooks.jlab.org/entry/3962660>. Submitted by tlemon<https://logbooks.jlab.org/user/tlemon> on Tue, 12/14/2021 - 08:54<https://logbooks.jlab.org/entries?start_date=1639486464&end_date=1639493664&book=DSGLOG&book=HBLOG&book=RICH>.

Logbooks:       DSGLOG<https://logbooks.jlab.org/book/dsglog> HBLOG<https://logbooks.jlab.org/book/hblog> RICH<https://logbooks.jlab.org/book/rich>
References:     3961560 - RICH is OFF due to the problems with the hardware interlock<https://logbooks.jlab.org/entry/3961560>

All modules in the hardware interlock system cRIO for the RICH electronic panel have been swapped (note: the cRIO chassis did not need to be swapped as the problem seemed to be with the modules).

All temperature and humidity sensors can now be seen in EPICS and the LabVIEW expert GUI.

The electronic panel hardware interlock system has now been restored to working order.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20211214/2f726e13/attachment.html>


More information about the dsg-hallb_rich mailing list