I agree,  we should switch to the oil performing this function. It may require some minor oil level adjustments when moved to the hall in order to prevent gas loss at this location.


--
George Jacobs
Jefferson Lab (TJNAF)
STE 12
12000 Jefferson Ave.
Newport News, VA 23606
(office) 757-269-7021
(email) jacobsg at jlab.org


________________________________
From: Dsg-rtpc <dsg-rtpc-bounces at jlab.org> on behalf of Sebastian Kuhn
Sent: Tuesday, June 11, 2019 4:46 PM
To: dsg-rtpc at jlab.org
Cc: Nathan Dzbenski; JIWAN POUDEL; Madhusudhan Pokhrel; Mohammad Hattawy
Subject: Re: [Dsg-rtpc] Gas panel Components

I am not saying that CV1 hasn’t been there all along, or that it didn’t have a purpose. I am just wondering whether we could remove it and instead rely on the bubbler to provide the overpressure relief. The advantage would be that it would engage already at much lower overpressure (for instance, we could have have one 1 cm of oil above the entrance tube, which would limit the overpressure to 10^-3 atmospheres or 0.014 psi. But it’s not a big issues either way...

On Jun 11, 2019, at 3:54 PM, George Jacobs <jacobsg at jlab.org<mailto:jacobsg at jlab.org>> wrote:

CV1 is a pressure bleed off valve. It's purpose is to limit pressure in the DMS gas volume to 1 psi in case of operator (human) error.


-------------- next part --------------
An HTML attachment was scrubbed...