[dsg-svt] DSG work request
Yuri Gotra
gotra at jlab.org
Tue Jul 31 09:11:10 EDT 2018
Follow up on the flow meter status.
- connectors disassembled, cleaned, contact to the pin opened, soldered
wire to the pin
- ordered 2 connector assemblies to replace corroded contacts
- flow meters mounted on the plate, tested, insulated
- insulation boxes sealed, purging lines with dry air installed
These actions will ensure flow meter protection against humidity.
On 7/17/18 11:37 AM, Marc McMullen wrote:
>
> Yuri,
>
> We have tested both flow meters, and noted inconsistencies in flow
> using multiple configurations. During testing this morning the "in"
> meter stopped working. We took the cover off and noted not only a
> moisture on the display cover, but standing water in the bottom of the
> enclosure.
>