[dsg-svt] Chiller Temperature

Brian Eng beng at jlab.org
Mon Sep 17 12:34:20 EDT 2018


Here's something cheap that seems like it should work?

https://amphenol-sensors.com/en/thermometrics/assemblies/3283-fluid-temperature-sensor

They're fairly cheap and still have a voltage output so it shouldn't be to hard to integrate with the existing system. Just need to order that plus the connector.

And actual RTD would have more range, but wouldn't be as straight forward to get into the controls system.

> On Sep 17, 2018, at 11:53 AM, Yuri Gotra <gotra at jlab.org> wrote:
> 
> Thank you Brian,
> 
> Yes, due to the incident with beam tuning the operating conditions of the SVT have been revised.
> 
> The flow range on this sensor was not right for the cooling system from the very beginning.
> 
> According to Nathan V450 has the following conversion: V=A*12.5/2^31.
> 
> and the temperature formula: T=V*15.2905 - 25.5352
> 
> I'll check with Bob what's available. The new chiller has adequate cooling capacity for current conditions and the bath temperature is close to the inlet coolant temperature.
> 
> 
> On 9/17/18 11:30 AM, Brian Eng wrote:
>> Hi Yuri,
>> 
>> According to the part number of the integrated flow/pressure/temperature sensor ( 0V7008FTPA32 ) it has a measurement range of -25 to 125 C because it has a voltage output for the temperature.
>> 
>> https://proteusind.com/wp-content/uploads/V7000DS.pdf
>> 
>> Since the operating parameters have changed so much (temperature, coolant, etc) for the chiller setup most of the sensors aren't really optimized for how it's currently configured.
>> 
>> Maybe see if there's some sensors that Hall B is already using that could be swapped in?