[Dsg-hallc_controls] Dipole Field Regulation PLC code completed
Brian Eng
beng at jlab.org
Thu Sep 27 12:41:34 EDT 2018
BTW it looks like it lost the lock around 2300 last night. The dipole was ramped down?
I had left some code in that gives up trying to get a lock after 6 failed attempts (now commented out).
However now it seems like the PT2026 won't get a lock even though the dipole is at field.
I've tried both their Windows software and their bundled LabVIEW VIs.
Not sure what the problem is because it looks like the field is at ~1.8 T which is well within the range of the probe.
----- Original Message -----
> From: "Brian Eng" <beng at jlab.org>
> To: "Mike Fowler" <fowler at jlab.org>
> Cc: "dsg-hallc controls" <dsg-hallc_controls at jlab.org>
> Sent: Wednesday, September 26, 2018 3:39:03 PM
> Subject: Re: [Dsg-hallc_controls] Dipole Field Regulation PLC code completed
> Short of connecting to that PLC, no.
>
> Should I also be writing to the controller tag Steve created on the HMS PLC as
> well?
>
> Since we don't have access to the HMI screens I don't know what the field the
> PT2025 is reporting (EPICS only has the dipole current), but currently the
> PT2026 has 1.4442039 T with the last 2 digits bouncing around, but it's staying
> pretty close to 40 (low of 36, high of 46 for the short time I was watching the
> tag).
>
> ----- Original Message -----
>> From: "Mike Fowler" <fowler at jlab.org>
>> To: "dsg-hallc controls" <dsg-hallc_controls at jlab.org>
>> Sent: Wednesday, September 26, 2018 2:55:38 PM
>> Subject: Re: [Dsg-hallc_controls] Dipole Field Regulation PLC code completed
>
>> On 9/26/2018 11:07 AM, Brian Eng wrote:
>>> the NMR is currently locked and sending field values
>>
>> Any way we can see the values?
>>
> > Mike
More information about the Dsg-hallc_controls
mailing list