> I wanted to let you know that we have begun analyzing the results of the
> stability tests with load.  I have attached a plot for one of the boards
> (#304) that detail the results of this test.
> As you can see, for this first board, all channels except for channels 11
> and 24 were within the manufacturer's specification of +/- 0.2 V. Channel 11
> was only ~120 mV below the lower range of accuracy and channel 24 was only
> ~50 mV below. The overall average voltage for all of the channels on Board
> #304 was 1499.92 which is ~80 mV lower than the expected value of 1500 V.
> We will of course put all of this information, and more, in a detailed
> report for you when we are done with our analysis.
> We'll continue to keep you informed of our progress. If you have any
> questions or concerns, please do not hesitate to contact us.
> Have a good weekend,
> Aaron
-- 