[Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin Question

Gianni Di Maio g.dimaio at caen.it
Mon Mar 2 12:12:26 EST 2020


Hello Aaron,

this may be can be a critic operation, and may be could induce the 
problem ( High Voltages discharges?), what do you think?
Depending from the grounding,  theHV  discharges could put in crisis 
some boards.
We use similar test set up with relays, to test the boards 
automatically, but we avoid to switch the relays with the presence of HV 
on the output.
Could you perform a test switching OFF the channels or putting the 
voltages at zero before switching the relays to verify if the problem 
persist?

Let me know.
Best regards.
Gianni

Il 02/03/2020 13:58, Aaron Brown ha scritto:
> Hello Gianni,
>
> Yes, that is correct. The relays are switched when there is high 
> voltages on the channels outputs.
>
> Best Regards,
> Aaron Brown
> ------------------------------------------------------------------------
> *From:* Gianni Di Maio <g.dimaio at caen.it>
> *Sent:* Monday, March 2, 2020 6:27 AM
> *To:* Aaron Brown <ambrown at jlab.org>
> *Cc:* Greg Kibilko <greg at setechsales.com>; dsg-hallc controls 
> <dsg-hallc_controls at jlab.org>; Alberto Lucchesi <a.lucchesi at caen.it>; 
> Claudio Landi <c.landi at caen.it>; a.orazzini at caen.it <a.orazzini at caen.it>
> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin 
> Question
> Dear Aaron,
>
> thank you for the details.
> It is not clear to me the sequence value (Vset) in the script you 
> sent,  when you say /"Test with load where the voltage was ramped up 
> step-wise. There were a series of relays that would quickly cycle 
> through all of the channels and a digital multi-meter would measure 
> the voltage output of the channel." /You mean that the relays are 
> switched when there is High voltages on the channels output?
>
> Best regards.
> Gianni
>
> Il 28/02/2020 20:09, Aaron Brown ha scritto:
>> Hello Gianni,
>>
>> I have attached pictures of the crates. They both have the boosters 
>> installed. Also, I'm attaching the script we used in GECO to run the 
>> test during which we experienced the board disconnection.
>> The script shows the voltages we were using for the test. There was 
>> no consistency in terms of losing connection to a board at a 
>> particular voltage. The disconnected boards never came back on their 
>> own, we have to power cycle the crate in order to re-establish 
>> communication with whichever board which we lost communication with.
>> Also, yesterday we experienced a board disconnection during a 
>> different test. It was during a ramping test with load (2 mega ohms) 
>> using GECO. This is the first time we've had this happen during this 
>> kind of test.
>> I have attached the script for this test as well.
>> Please do not hesitate to reach out if you have any additional questions.
>>
>> Best Regards,
>> Aaron Brown
>> ------------------------------------------------------------------------
>> *From:* Gianni Di Maio <g.dimaio at caen.it> <mailto:g.dimaio at caen.it>
>> *Sent:* Thursday, February 27, 2020 12:12 PM
>> *To:* Aaron Brown <ambrown at jlab.org> <mailto:ambrown at jlab.org>
>> *Cc:* Greg Kibilko <greg at setechsales.com> 
>> <mailto:greg at setechsales.com>; dsg-hallc controls 
>> <dsg-hallc_controls at jlab.org> <mailto:dsg-hallc_controls at jlab.org>; 
>> Alberto Lucchesi <a.lucchesi at caen.it> <mailto:a.lucchesi at caen.it>; 
>> Claudio Landi <c.landi at caen.it> <mailto:c.landi at caen.it>
>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin 
>> Question
>> Dear Aaron,
>>
>> thank you very much for the detailed report.
>> I have some questions for you.
>> The Main Frame that host 16  A7030TN modules how is equipped? Do you 
>> have at least one  48V booster low voltage power supply  plugged in?
>> Could you send me a picture of the Main Frame front panel?
>> When a board or more boards that populated the crate disappear, after 
>> a while it re-appear or are not present any more in the crate map? In 
>> this case you have to do a power cycle to see again the boards that 
>> disappear?
>> I had a look to the HV Module Test Stand Setup circuit, I would like 
>> to know if when you switch the relays, it is present the HV on the 
>> channel, if yes at what voltage you perform the test?
>>
>> Let me know.
>>
>> Best regards.
>>
>> Gianni
>>
>>
>> Il 27/02/2020 15:27, Greg Kibilko ha scritto:
>>> Hi Aaron - Thank you very much.
>>>
>>> *Ciao Alberto and Gianni - If you are not currently duplicating this 
>>> setup then perhaps this information may help you replicate and/or 
>>> better understand the root cause of these disconnects.**
>>> *
>>> Kind regards,
>>> Greg
>>>
>>> On 2/27/2020 8:13 AM, Aaron Brown wrote:
>>>> Hello Greg,
>>>>
>>>> So, let me try to explain what was going on when we noticed the 
>>>> random disconnections.
>>>> We were conducting a test with load where the voltage was ramped up 
>>>> step-wise. There were a series of relays that would quickly cycle 
>>>> through all of the channels and a digital multi-meter would measure 
>>>> the voltage output of the channel.
>>>> The disconnection occurred on the board attached to the load, and 
>>>> sometimes to other random boards populated in the crate.
>>>> I have attached a talk given by one of our group members. Pages 
>>>> 12-14 detail the test setup.
>>>> Let me know if you have any other questions.
>>>> Hopefully this will be helpful.
>>>>
>>>> Best regards,
>>>> Aaron Brown
>>>> ------------------------------------------------------------------------
>>>> *From:* Greg Kibilko <greg at setechsales.com> 
>>>> <mailto:greg at setechsales.com>
>>>> *Sent:* Tuesday, February 25, 2020 9:21 AM
>>>> *To:* Aaron Brown <ambrown at jlab.org> <mailto:ambrown at jlab.org>
>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin 
>>>> Question
>>>> Thank you sir. Alberto's diagnosis on the source of the parameter 
>>>> change problem is heartening, but we're just trying to understand 
>>>> why the other problem (random disconnect) is still not being seen 
>>>> at CAEN.
>>>>
>>>> Cheers,
>>>> Greg
>>>>
>>>> On 2/25/2020 9:19 AM, Aaron Brown wrote:
>>>>> Good morning Greg,
>>>>>
>>>>> To my knowledge there is no correlation between the pin issue and 
>>>>> the disconnects.
>>>>> I will ask around if anyone has noticed a connection between these 
>>>>> two issues, but it is my understanding that they are separate.
>>>>>
>>>>> Best Regards,
>>>>> Aaron Brown
>>>>> ------------------------------------------------------------------------
>>>>> *From:* Greg Kibilko <greg at setechsales.com> 
>>>>> <mailto:greg at setechsales.com>
>>>>> *Sent:* Tuesday, February 25, 2020 8:28 AM
>>>>> *To:* Aaron Brown <ambrown at jlab.org> <mailto:ambrown at jlab.org>; 
>>>>> dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>>>> <mailto:dsg-hallc_controls at jlab.org>
>>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin 
>>>>> Question
>>>>> Hi Aaron,
>>>>>
>>>>> I'm curious if you've seen any correlation between the disconnects 
>>>>> and the problems with the boards that have the SL pin pushing out? 
>>>>> Could there be a connection?
>>>>>
>>>>> Best,
>>>>> Greg
>>>>>
>>>>> On 2/24/2020 11:38 AM, Aaron Brown wrote:
>>>>>> Hello Alberto,
>>>>>>
>>>>>> This is wonderful news!
>>>>>> I look forward to receiving the new HiVocs version soon.
>>>>>>
>>>>>> Best Regards,
>>>>>> Aaron Brown
>>>>>> ------------------------------------------------------------------------
>>>>>> *From:* Alberto Lucchesi <a.lucchesi at caen.it> 
>>>>>> <mailto:a.lucchesi at caen.it>
>>>>>> *Sent:* Monday, February 24, 2020 10:22 AM
>>>>>> *To:* Aaron Brown <ambrown at jlab.org> <mailto:ambrown at jlab.org>; 
>>>>>> Gianni Di Maio <g.dimaio at caen.it> <mailto:g.dimaio at caen.it>; Greg 
>>>>>> Kibilko <greg at setechsales.com> <mailto:greg at setechsales.com>; 
>>>>>> Amrit Yegneswaran <yeg at jlab.org> <mailto:yeg at jlab.org>
>>>>>> *Cc:* Patrizia Rossi <rossi at jlab.org> <mailto:rossi at jlab.org>; 
>>>>>> dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>>>>> <mailto:dsg-hallc_controls at jlab.org>
>>>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin 
>>>>>> Question
>>>>>> Hi Aaron,
>>>>>>
>>>>>> I'was able to reproduce the random parameter changes issue.
>>>>>>
>>>>>> The issue was not visible with only one board because it is 
>>>>>> related to the parallelism inside the IOC.
>>>>>>
>>>>>> What was happening is that with several boards the EPICS 
>>>>>> framework (our IOC is based over it) calls parts of our code in 
>>>>>> parallel and there was an internal structure that was not 
>>>>>> thread-safe. The results were that some settings were not  
>>>>>> applied to the right parameter but to onother one (the aren't 
>>>>>> random parameter changes, The values were right but the 
>>>>>> parameters were wrong! (i.e., a Trip = 40 were codified as VSet = 
>>>>>> 40 or a PW = On where codified as RUp = 1).
>>>>>>
>>>>>> I fixed the issue and now I'm testing the results and the issues 
>>>>>> seem to be resolved. I'll do other tests in these days then I'll 
>>>>>> send you a new HiVocs version within a couple of days.
>>>>>>
>>>>>> I don't have news about the other issue (the board disconnection) 
>>>>>> in all my test I never experienced a board disconnection.
>>>>>>
>>>>>>
>>>>>> Best Regards,
>>>>>> Albert
>>>>>>
>>>>>>
>>>>>> Il 24/02/2020 14:14, Aaron Brown ha scritto:
>>>>>>> Hello Gianni,
>>>>>>>
>>>>>>> Yes, April works for us. We will have all the boards in our 
>>>>>>> possession on hand for any rework that can be done.
>>>>>>>
>>>>>>> Best regards,
>>>>>>> Aaron Brown
>>>>>>> ------------------------------------------------------------------------
>>>>>>> *From:* Gianni Di Maio <g.dimaio at caen.it> <mailto:g.dimaio at caen.it>
>>>>>>> *Sent:* Friday, February 21, 2020 5:13 PM
>>>>>>> *To:* Greg Kibilko <greg at setechsales.com> 
>>>>>>> <mailto:greg at setechsales.com>; Aaron Brown <ambrown at jlab.org> 
>>>>>>> <mailto:ambrown at jlab.org>; Amrit Yegneswaran <yeg at jlab.org> 
>>>>>>> <mailto:yeg at jlab.org>
>>>>>>> *Cc:* Patrizia Rossi <rossi at jlab.org> <mailto:rossi at jlab.org>; 
>>>>>>> dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>>>>>> <mailto:dsg-hallc_controls at jlab.org>; Alberto Lucchesi 
>>>>>>> <a.lucchesi at caen.it> <mailto:a.lucchesi at caen.it>
>>>>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector 
>>>>>>> Pin Question
>>>>>>> Dear Aaron,
>>>>>>>
>>>>>>> as told by Greg April could be OK with us to visit you.
>>>>>>> If possible I would like to have a look to all the boards 
>>>>>>> connectors ( to rework partially the RADIALL connector and in 
>>>>>>> particularly the safety loop pin).
>>>>>>> Anyway for the moment just let me know if April is OK,  I will 
>>>>>>> send later a list of rework I would like to do on the boards and 
>>>>>>> cables.
>>>>>>>
>>>>>>> Have a nice week end to everybody
>>>>>>> Gianni
>>>>>>> Il 18/02/2020 21:03, Greg Kibilko ha scritto:
>>>>>>>> Hi Aaron,
>>>>>>>>
>>>>>>>> Very good.  Gianni is home sick, but I did Skype with him and 
>>>>>>>> his schedule for March is full.  We're aiming for April.  I 
>>>>>>>> should be able to propose some dates for you this week.
>>>>>>>>
>>>>>>>> Best,
>>>>>>>> Greg
>>>>>>>>
>>>>>>>> On 2/18/2020 1:31 PM, Aaron Brown wrote:
>>>>>>>>> Hi Greg,
>>>>>>>>>
>>>>>>>>> If April is the best time for you guys, it works for us.
>>>>>>>>>
>>>>>>>>> Best regards,
>>>>>>>>> Aaron
>>>>>>>>> ------------------------------------------------------------------------
>>>>>>>>> *From:* Greg Kibilko <greg at setechsales.com> 
>>>>>>>>> <mailto:greg at setechsales.com>
>>>>>>>>> *Sent:* Tuesday, February 18, 2020 10:38 AM
>>>>>>>>> *To:* Aaron Brown <ambrown at jlab.org> 
>>>>>>>>> <mailto:ambrown at jlab.org>; Gianni Di Maio <g.dimaio at caen.it> 
>>>>>>>>> <mailto:g.dimaio at caen.it>; Amrit Yegneswaran <yeg at jlab.org> 
>>>>>>>>> <mailto:yeg at jlab.org>
>>>>>>>>> *Cc:* Patrizia Rossi <rossi at jlab.org> <mailto:rossi at jlab.org>; 
>>>>>>>>> dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>>>>>>>> <mailto:dsg-hallc_controls at jlab.org>; Alberto Lucchesi 
>>>>>>>>> <a.lucchesi at caen.it> <mailto:a.lucchesi at caen.it>
>>>>>>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector 
>>>>>>>>> Pin Question
>>>>>>>>> Hi Aaron,
>>>>>>>>>
>>>>>>>>> Ok, fantastic.  This gives us plenty of time to work with.
>>>>>>>>>
>>>>>>>>> How about timeline?  Is April too late?
>>>>>>>>>
>>>>>>>>> Cheers,
>>>>>>>>> Greg
>>>>>>>>>
>>>>>>>>> On 2/18/2020 10:36 AM, Aaron Brown wrote:
>>>>>>>>>> Morning Greg,
>>>>>>>>>>
>>>>>>>>>> The only day/time I can think of that /might/ pose a problem 
>>>>>>>>>> would be Wednesday afternoons between 1 and 3.
>>>>>>>>>> Other than that, there is nothing that can't be worked around.
>>>>>>>>>>
>>>>>>>>>> Thanks,
>>>>>>>>>> Aaron
>>>>>>>>>> ------------------------------------------------------------------------
>>>>>>>>>> *From:* Greg Kibilko <greg at setechsales.com> 
>>>>>>>>>> <mailto:greg at setechsales.com>
>>>>>>>>>> *Sent:* Tuesday, February 18, 2020 10:32 AM
>>>>>>>>>> *To:* Aaron Brown <ambrown at jlab.org> 
>>>>>>>>>> <mailto:ambrown at jlab.org>; Gianni Di Maio <g.dimaio at caen.it> 
>>>>>>>>>> <mailto:g.dimaio at caen.it>; Amrit Yegneswaran <yeg at jlab.org> 
>>>>>>>>>> <mailto:yeg at jlab.org>
>>>>>>>>>> *Cc:* Patrizia Rossi <rossi at jlab.org> 
>>>>>>>>>> <mailto:rossi at jlab.org>; dsg-hallc controls 
>>>>>>>>>> <dsg-hallc_controls at jlab.org> 
>>>>>>>>>> <mailto:dsg-hallc_controls at jlab.org>; Alberto Lucchesi 
>>>>>>>>>> <a.lucchesi at caen.it> <mailto:a.lucchesi at caen.it>
>>>>>>>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector 
>>>>>>>>>> Pin Question
>>>>>>>>>> Hi Aaron,
>>>>>>>>>>
>>>>>>>>>> I am sure Gianni will agree that a visit is in order.
>>>>>>>>>>
>>>>>>>>>> In the next couple of months are there any specific 
>>>>>>>>>> days/weeks we should target to visit?  Or days/weeks we 
>>>>>>>>>> should try to avoid?
>>>>>>>>>>
>>>>>>>>>> Thanks!
>>>>>>>>>> Greg
>>>>>>>>>>
>>>>>>>>>> On 2/18/2020 10:30 AM, Aaron Brown wrote:
>>>>>>>>>>> Hello Gianni,
>>>>>>>>>>>
>>>>>>>>>>> We currently have 33 A7030TN modules and 2 SY4527 crates on 
>>>>>>>>>>> hand.
>>>>>>>>>>> I think that best idea would be to have someone from CAEN 
>>>>>>>>>>> make a visit to the lab and take a look/fix the connector pins.
>>>>>>>>>>>
>>>>>>>>>>> Best regards,
>>>>>>>>>>> Aaron Brown
>>>>>>>>>>> ------------------------------------------------------------------------
>>>>>>>>>>> *From:* Gianni Di Maio <g.dimaio at caen.it> 
>>>>>>>>>>> <mailto:g.dimaio at caen.it>
>>>>>>>>>>> *Sent:* Monday, February 17, 2020 12:12 PM
>>>>>>>>>>> *To:* Aaron Brown <ambrown at jlab.org> 
>>>>>>>>>>> <mailto:ambrown at jlab.org>; Amrit Yegneswaran <yeg at jlab.org> 
>>>>>>>>>>> <mailto:yeg at jlab.org>
>>>>>>>>>>> *Cc:* Patrizia Rossi <rossi at jlab.org> 
>>>>>>>>>>> <mailto:rossi at jlab.org>; dsg-hallc controls 
>>>>>>>>>>> <dsg-hallc_controls at jlab.org> 
>>>>>>>>>>> <mailto:dsg-hallc_controls at jlab.org>; Greg Kibilko 
>>>>>>>>>>> <greg at setechsales.com> <mailto:greg at setechsales.com>; 
>>>>>>>>>>> Alberto Lucchesi <a.lucchesi at caen.it> 
>>>>>>>>>>> <mailto:a.lucchesi at caen.it>
>>>>>>>>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN 
>>>>>>>>>>> Connector Pin Question
>>>>>>>>>>> Hello Aaron,
>>>>>>>>>>> thank you very much for the details.
>>>>>>>>>>> Really strange a so high number of problems on the panel 
>>>>>>>>>>> connector, as told from our experience it is more 
>>>>>>>>>>> problematic the cable connector, this is the reason that we 
>>>>>>>>>>> add the hot glue.
>>>>>>>>>>> At this point let me know how you wish proceed.
>>>>>>>>>>> To rework/repair the boards with pins pushed out we have two 
>>>>>>>>>>> solutions, you ship the boards to our office in NY, or I 
>>>>>>>>>>> could make an intervention in JLAB in spring.
>>>>>>>>>>> How many boards you have in total?
>>>>>>>>>>> May be it's worth to have a look to al the boards.
>>>>>>>>>>>
>>>>>>>>>>> Waiting your comments.
>>>>>>>>>>> Best regards.
>>>>>>>>>>> Gianni
>>>>>>>>>>>
>>>>>>>>>>> Il 14/02/2020 16:48, Aaron Brown ha scritto:
>>>>>>>>>>>> Hello Gianni,
>>>>>>>>>>>>
>>>>>>>>>>>> We have experienced the issue of pins getting pushed back 
>>>>>>>>>>>> 12 separate times;
>>>>>>>>>>>> 8 times it was the pins for the safety loops, and 4 times 
>>>>>>>>>>>> is was an arbitrary pin.
>>>>>>>>>>>> All occurrences were on the board R52 connector (male).
>>>>>>>>>>>> I have attached a photo to highlight exactly what we're 
>>>>>>>>>>>> describing.
>>>>>>>>>>>> Also attached is a photo of the inside of one of the boards 
>>>>>>>>>>>> showing how the channel pins
>>>>>>>>>>>> are strain relieved, but the safety loop pins are not.
>>>>>>>>>>>>
>>>>>>>>>>>> If you have any questions, please do not hesitate to send 
>>>>>>>>>>>> an email to the mailing list.
>>>>>>>>>>>>
>>>>>>>>>>>> Warm regards,
>>>>>>>>>>>> Aaron Brown
>>>>>>>>>>>> ------------------------------------------------------------------------
>>>>>>>>>>>> *From:* Dsg-hallc_controls 
>>>>>>>>>>>> <dsg-hallc_controls-bounces at jlab.org> 
>>>>>>>>>>>> <mailto:dsg-hallc_controls-bounces at jlab.org> on behalf of 
>>>>>>>>>>>> Gianni Di Maio <g.dimaio at caen.it> <mailto:g.dimaio at caen.it>
>>>>>>>>>>>> *Sent:* Friday, February 14, 2020 3:26 AM
>>>>>>>>>>>> *To:* Amrit Yegneswaran <yeg at jlab.org> <mailto:yeg at jlab.org>
>>>>>>>>>>>> *Cc:* Patrizia Rossi <rossi at jlab.org> 
>>>>>>>>>>>> <mailto:rossi at jlab.org>; dsg-hallc controls 
>>>>>>>>>>>> <dsg-hallc_controls at jlab.org> 
>>>>>>>>>>>> <mailto:dsg-hallc_controls at jlab.org>; Greg Kibilko 
>>>>>>>>>>>> <greg at setechsales.com> <mailto:greg at setechsales.com>; 
>>>>>>>>>>>> Alberto Lucchesi <a.lucchesi at caen.it> 
>>>>>>>>>>>> <mailto:a.lucchesi at caen.it>
>>>>>>>>>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN 
>>>>>>>>>>>> Connector Pin Question
>>>>>>>>>>>> Dear Amrit,
>>>>>>>>>>>>
>>>>>>>>>>>> thank you for your email, I apologize I was not aware about 
>>>>>>>>>>>> the dsg mailing list (dsg-hallc_controls at jlab.org 
>>>>>>>>>>>> <mailto:dsg-hallc_controls at jlab.org>).
>>>>>>>>>>>> I full agree with you the pushed pins issue will be surely 
>>>>>>>>>>>> addressed by CAEN, we are considering to do a maintenance 
>>>>>>>>>>>> intervention at JLAB in the next future.
>>>>>>>>>>>> What I would like to know  to have a clear idea of the 
>>>>>>>>>>>> situation is
>>>>>>>>>>>> How many boards with pushed pin/s you have?
>>>>>>>>>>>> How many pins per board?
>>>>>>>>>>>> If I well understood ( but may be I'm wrong) there are 4 
>>>>>>>>>>>> boards with this issue, that is one pin pushed back,and 
>>>>>>>>>>>> every time is one of the two safety pin loop.
>>>>>>>>>>>>
>>>>>>>>>>>> Could you or some one of the group confirm me these numbers?
>>>>>>>>>>>>
>>>>>>>>>>>> Or in case I'm wrong could you provide me a report updated 
>>>>>>>>>>>> to today?
>>>>>>>>>>>>
>>>>>>>>>>>>
>>>>>>>>>>>> Best regards.
>>>>>>>>>>>>
>>>>>>>>>>>> Gianni
>>>>>>>>>>>>
>>>>>>>>>>>>
>>>>>>>>>>>> Il 13/02/2020 19:16, Amrit Yegneswaran ha scritto:
>>>>>>>>>>>>> hello,
>>>>>>>>>>>>> this issue of pushed pins needs to be addressed by CAEN.
>>>>>>>>>>>>> dsg is not in this business.
>>>>>>>>>>>>> also, i am very disappointed that i'm receiving the 
>>>>>>>>>>>>> gianni's e-mail only today now.
>>>>>>>>>>>>> i urge all of you, for the sake of transparency, 
>>>>>>>>>>>>> particularly from the caen side to use the mailing list, 
>>>>>>>>>>>>> like we do here at jlab.
>>>>>>>>>>>>> amrit
>>>>>>>>>>>>> if anyone in the dsg group receives a "singular" e-mail 
>>>>>>>>>>>>> make sure you broadcast it to the mailing list.
>>>>>>>>>>>>> ------------------------------------------------------------------------
>>>>>>>>>>>>> *From:* Dsg-hallc_controls 
>>>>>>>>>>>>> <dsg-hallc_controls-bounces at jlab.org> 
>>>>>>>>>>>>> <mailto:dsg-hallc_controls-bounces at jlab.org> on behalf of 
>>>>>>>>>>>>> Brian Eng <beng at jlab.org> <mailto:beng at jlab.org>
>>>>>>>>>>>>> *Sent:* Thursday, February 13, 2020 12:32 PM
>>>>>>>>>>>>> *To:* dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>>>>>>>>>>>> <mailto:dsg-hallc_controls at jlab.org>
>>>>>>>>>>>>> *Subject:* [Dsg-hallc_controls] Fwd: [EXTERNAL] CAEN 
>>>>>>>>>>>>> Connector Pin Question
>>>>>>>>>>>>> In regards to Brad's previous email about the potting 
>>>>>>>>>>>>> material here's one of the emails CAEN sent.
>>>>>>>>>>>>>
>>>>>>>>>>>>> Still waiting on more feedback since we've definitely seen 
>>>>>>>>>>>>> it on the board side multiple times.
>>>>>>>>>>>>>
>>>>>>>>>>>>>> Begin forwarded message:
>>>>>>>>>>>>>>
>>>>>>>>>>>>>> *From: *Gianni Di Maio <g.dimaio at caen.it 
>>>>>>>>>>>>>> <mailto:g.dimaio at caen.it>>
>>>>>>>>>>>>>> *Subject: **Re: [EXTERNAL] CAEN Connector Pin Question*
>>>>>>>>>>>>>> *Date: *February 6, 2020 at 5:42:31 PM EST
>>>>>>>>>>>>>> *To: *Brian Eng <beng at jlab.org <mailto:beng at jlab.org>>
>>>>>>>>>>>>>> *Cc: *Greg Kibilko <greg at setechsales.com 
>>>>>>>>>>>>>> <mailto:greg at setechsales.com>>
>>>>>>>>>>>>>>
>>>>>>>>>>>>>> Dear Brian,
>>>>>>>>>>>>>>
>>>>>>>>>>>>>> please could you tell me if the pins pushed out are 
>>>>>>>>>>>>>> present only on the RADIALL connector installed on the 
>>>>>>>>>>>>>> boards (female)  or even on the RADIALL connectors 
>>>>>>>>>>>>>> installed on the cables ( male).
>>>>>>>>>>>>>> >From my experience usually in the past the problem 
>>>>>>>>>>>>>> appeared only on cable side, this is the reason that we 
>>>>>>>>>>>>>> decided to add the hot glue, in order to strengthen the 
>>>>>>>>>>>>>> tightness of the pin and prevent a possible escape from 
>>>>>>>>>>>>>> the correct position, but this as said only for the cable 
>>>>>>>>>>>>>> connector.
>>>>>>>>>>>>>> The problem on the panel connector is very rare, how many 
>>>>>>>>>>>>>> problematic boards you have  with pins pushed out?
>>>>>>>>>>>>>> Could you send me some pictures?
>>>>>>>>>>>>>>
>>>>>>>>>>>>>> Best regards.
>>>>>>>>>>>>>> Gianni
>>>>>>>>>>>>>
>>>>>>>>>>>>
>>>>>>>>>>>> -- 
>>>>>>>>>>>>
>>>>>>>>>>>>  CAEN S.p.A.  Gianni Di Maio
>>>>>>>>>>>>
>>>>>>>>>>>> *Maintenance Division Manager*
>>>>>>>>>>>>
>>>>>>>>>>>> Email: g.dimaio at caen.it <mailto:g.dimaio at caen.it>
>>>>>>>>>>>>
>>>>>>>>>>>> Skype: gdimaio_caen
>>>>>>>>>>>>
>>>>>>>>>>>> 	Cell:+393482556525 	
>>>>>>>>>>>>
>>>>>>>>>>>> *CAEN S.p.A.*
>>>>>>>>>>>>
>>>>>>>>>>>> Tel.: +39 0584 388 398
>>>>>>>>>>>> Fax: +39 0584 388 959
>>>>>>>>>>>> Email: info at caen.it <mailto:info at caen.it>
>>>>>>>>>>>> via Vetraia,11 • 55049 • Viareggio (LU) • Italy
>>>>>>>>>>>> Cod.Fisc. e p.IVA IT00864500467
>>>>>>>>>>>> https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwIF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=Y_6u1GUQugVDYN3xLTG55vJvalWvRQDGwLbMtQADg4g&s=v2iX4fTA7dIEEaBAAVjieKWdSnGj_t_6h3WZCaNfS-0&e=  
>>>>>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=NkBWvHI5Zs-9oXlFaDnzpnLIvyQVuV_qo4jAGvl6JM8&s=HXuJxESZtIm45hzrZH6q0YrYEYtV0J5kDumRkl8BGW4&e=> 
>>>>>>>>>>>>
>>>>>>>>>>>>
>>>>>>>>>>>> Le informazioni trasmesse attraverso la presente 
>>>>>>>>>>>> comunicazione via mail ed eventuali allegati sono da 
>>>>>>>>>>>> intendersi di esclusiva spettanza del destinatario. Nel 
>>>>>>>>>>>> caso in cui le stesse raggiungessero, per qualunque motivo, 
>>>>>>>>>>>> soggetti diversi dai destinatari, questi dovrebbero darne 
>>>>>>>>>>>> immediata notizia al mittente. In ogni caso, eventuali 
>>>>>>>>>>>> soggetti diversi dai legittimi destinatari della presente 
>>>>>>>>>>>> comunicazione sono esplicitamente diffidati da ogni 
>>>>>>>>>>>> utilizzazione, sia ai sensi del D.Lgs. n. 196/2003 "Codice 
>>>>>>>>>>>> in materia di protezione dei dati personali" sia ai sensi 
>>>>>>>>>>>> degli art. 616 e ss. Codice Penale che sanziona la 
>>>>>>>>>>>> violazione del segreto sulla corrispondenza.
>>>>>>>>>>>> The information contained in this transmission may contain 
>>>>>>>>>>>> privileged and confidential information and is intended 
>>>>>>>>>>>> only for the use of the person(s)named above. If you are 
>>>>>>>>>>>> not the intended recipient, or an employee or agent 
>>>>>>>>>>>> responsible for delivering this message to the intended 
>>>>>>>>>>>> recipient, any review, dissemination, distribution or 
>>>>>>>>>>>> duplication of this communication is strictly prohibited. 
>>>>>>>>>>>> If you are not the intended recipient, please contact the 
>>>>>>>>>>>> sender immediately and destroy all copies of the original 
>>>>>>>>>>>> message.
>>>>>>>>>>>>
>>>>>>>>>>>
>>>>>>>>>>> -- 
>>>>>>>>>>>
>>>>>>>>>>>  CAEN S.p.A.  Gianni Di Maio
>>>>>>>>>>>
>>>>>>>>>>> *Maintenance Division Manager*
>>>>>>>>>>>
>>>>>>>>>>> Email: g.dimaio at caen.it <mailto:g.dimaio at caen.it>
>>>>>>>>>>>
>>>>>>>>>>> Skype: gdimaio_caen
>>>>>>>>>>>
>>>>>>>>>>> 	Cell:+393482556525 	
>>>>>>>>>>>
>>>>>>>>>>> *CAEN S.p.A.*
>>>>>>>>>>>
>>>>>>>>>>> Tel.: +39 0584 388 398
>>>>>>>>>>> Fax: +39 0584 388 959
>>>>>>>>>>> Email: info at caen.it <mailto:info at caen.it>
>>>>>>>>>>> via Vetraia,11 • 55049 • Viareggio (LU) • Italy
>>>>>>>>>>> Cod.Fisc. e p.IVA IT00864500467
>>>>>>>>>>> https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwIF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=Y_6u1GUQugVDYN3xLTG55vJvalWvRQDGwLbMtQADg4g&s=v2iX4fTA7dIEEaBAAVjieKWdSnGj_t_6h3WZCaNfS-0&e=  
>>>>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=CK4IA3vcXcfmGx6m3v3SLbVbALDNmIRVqwR42dIp0vQ&s=3g8vV8eb0W81I5E9dK0wOFl7hjtwyjGD3KSLHzdqEd4&e=> 
>>>>>>>>>>>
>>>>>>>>>>>
>>>>>>>>>>> Le informazioni trasmesse attraverso la presente 
>>>>>>>>>>> comunicazione via mail ed eventuali allegati sono da 
>>>>>>>>>>> intendersi di esclusiva spettanza del destinatario. Nel caso 
>>>>>>>>>>> in cui le stesse raggiungessero, per qualunque motivo, 
>>>>>>>>>>> soggetti diversi dai destinatari, questi dovrebbero darne 
>>>>>>>>>>> immediata notizia al mittente. In ogni caso, eventuali 
>>>>>>>>>>> soggetti diversi dai legittimi destinatari della presente 
>>>>>>>>>>> comunicazione sono esplicitamente diffidati da ogni 
>>>>>>>>>>> utilizzazione, sia ai sensi del D.Lgs. n. 196/2003 "Codice 
>>>>>>>>>>> in materia di protezione dei dati personali" sia ai sensi 
>>>>>>>>>>> degli art. 616 e ss. Codice Penale che sanziona la 
>>>>>>>>>>> violazione del segreto sulla corrispondenza.
>>>>>>>>>>> The information contained in this transmission may contain 
>>>>>>>>>>> privileged and confidential information and is intended only 
>>>>>>>>>>> for the use of the person(s)named above. If you are not the 
>>>>>>>>>>> intended recipient, or an employee or agent responsible for 
>>>>>>>>>>> delivering this message to the intended recipient, any 
>>>>>>>>>>> review, dissemination, distribution or duplication of this 
>>>>>>>>>>> communication is strictly prohibited. If you are not the 
>>>>>>>>>>> intended recipient, please contact the sender immediately 
>>>>>>>>>>> and destroy all copies of the original message.
>>>>>>>>>>>
>>>>>>>>>>
>>>>>>>>>> -- 
>>>>>>>>>> *Greg Kibilko*
>>>>>>>>>> *SETS - SE Technical Sales, Inc.*
>>>>>>>>>> 112 Tyra Lane
>>>>>>>>>> Townville, SC 29689
>>>>>>>>>> P: (864) 356-2606
>>>>>>>>>> CAEN Products 
>>>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=Ny4G7qu4sv_T0NtiToDEcRuwjKB8XwFEyYHHIK3W5Sg&s=3nJ2zMdxx2iYrQfVevkoQ136crvz5WPX-u9vLXbDKyc&e=>
>>>>>>>>>> CAEN SyS Products & Solutions 
>>>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=Ny4G7qu4sv_T0NtiToDEcRuwjKB8XwFEyYHHIK3W5Sg&s=skGqBl6x9y3h92KiLpckUYhVQAtLyVD90zuz4wljKO4&e=>
>>>>>>>>>> IOxOs Products 
>>>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=Ny4G7qu4sv_T0NtiToDEcRuwjKB8XwFEyYHHIK3W5Sg&s=OLAIrxGqeN6aRvKctgEs_ax0G-qYK9hWu6Do-hdBywM&e=>
>>>>>>>>>
>>>>>>>>> -- 
>>>>>>>>> *Greg Kibilko*
>>>>>>>>> *SETS - SE Technical Sales, Inc.*
>>>>>>>>> 112 Tyra Lane
>>>>>>>>> Townville, SC 29689
>>>>>>>>> P: (864) 356-2606
>>>>>>>>> CAEN Products 
>>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=RuKyQ246VCfLgDisBhNfXaVwXLPxS8snA1X68m1DNYk&s=OnPfX_7lx0ht8FdYXuDrd7OFUQ58Ifa1FEn5qe97ndk&e=>
>>>>>>>>> CAEN SyS Products & Solutions 
>>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=RuKyQ246VCfLgDisBhNfXaVwXLPxS8snA1X68m1DNYk&s=LjH-sZKE_oLptW3aG24eGtFrP1pCGT8A5-PZU5mC3xI&e=>
>>>>>>>>> IOxOs Products 
>>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=RuKyQ246VCfLgDisBhNfXaVwXLPxS8snA1X68m1DNYk&s=cxNHULDWIflLFm2_PSTqcGcmRiirjlGJyPqqHNTYsB8&e=>
>>>>>>>>
>>>>>>>> -- 
>>>>>>>> *Greg Kibilko*
>>>>>>>> *SETS - SE Technical Sales, Inc.*
>>>>>>>> 112 Tyra Lane
>>>>>>>> Townville, SC 29689
>>>>>>>> P: (864) 356-2606
>>>>>>>> CAEN Products 
>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=ksMkvqymFO1sFCHsTbtFDMSRRxzdO2ZNaUaJOP8I-po&s=wWIxhZ0sQFRqCM2a7qnyYSru8vh21i-yFlS9DiglFLE&e=>
>>>>>>>> CAEN SyS Products & Solutions 
>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=ksMkvqymFO1sFCHsTbtFDMSRRxzdO2ZNaUaJOP8I-po&s=zcy1CKSquMtAvMb6ArLv6fcDY94k9m3jXSVKz6kgGrQ&e=>
>>>>>>>> IOxOs Products 
>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=ksMkvqymFO1sFCHsTbtFDMSRRxzdO2ZNaUaJOP8I-po&s=YJAckWJYg3im1sjAx40VNIKTFS7CLO9TnLUALivkW6M&e=>
>>>>>>>
>>>>>>>
>>>>>>> -- 
>>>>>>>
>>>>>>>  CAEN S.p.A.  Gianni Di Maio
>>>>>>>
>>>>>>> *Maintenance Division Manager*
>>>>>>>
>>>>>>> Email: g.dimaio at caen.it <mailto:g.dimaio at caen.it>
>>>>>>>
>>>>>>> Skype: gdimaio_caen
>>>>>>>
>>>>>>> 	Cell:+393482556525 	
>>>>>>>
>>>>>>> *CAEN S.p.A.*
>>>>>>>
>>>>>>> Tel.: +39 0584 388 398
>>>>>>> Fax: +39 0584 388 959
>>>>>>> Email: info at caen.it <mailto:info at caen.it>
>>>>>>> via Vetraia,11 • 55049 • Viareggio (LU) • Italy
>>>>>>> Cod.Fisc. e p.IVA IT00864500467
>>>>>>> https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwIF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=Y_6u1GUQugVDYN3xLTG55vJvalWvRQDGwLbMtQADg4g&s=v2iX4fTA7dIEEaBAAVjieKWdSnGj_t_6h3WZCaNfS-0&e=  
>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=ksMkvqymFO1sFCHsTbtFDMSRRxzdO2ZNaUaJOP8I-po&s=m_dmwoWv0btIaKtYcLCUaVxX_F_Gy3P48QiYGL-j4Cw&e=> 
>>>>>>>
>>>>>>>
>>>>>>> Le informazioni trasmesse attraverso la presente comunicazione 
>>>>>>> via mail ed eventuali allegati sono da intendersi di esclusiva 
>>>>>>> spettanza del destinatario. Nel caso in cui le stesse 
>>>>>>> raggiungessero, per qualunque motivo, soggetti diversi dai 
>>>>>>> destinatari, questi dovrebbero darne immediata notizia al 
>>>>>>> mittente. In ogni caso, eventuali soggetti diversi dai legittimi 
>>>>>>> destinatari della presente comunicazione sono esplicitamente 
>>>>>>> diffidati da ogni utilizzazione, sia ai sensi del D.Lgs. n. 
>>>>>>> 196/2003 "Codice in materia di protezione dei dati personali" 
>>>>>>> sia ai sensi degli art. 616 e ss. Codice Penale che sanziona la 
>>>>>>> violazione del segreto sulla corrispondenza.
>>>>>>> The information contained in this transmission may contain 
>>>>>>> privileged and confidential information and is intended only for 
>>>>>>> the use of the person(s)named above. If you are not the intended 
>>>>>>> recipient, or an employee or agent responsible for delivering 
>>>>>>> this message to the intended recipient, any review, 
>>>>>>> dissemination, distribution or duplication of this communication 
>>>>>>> is strictly prohibited. If you are not the intended recipient, 
>>>>>>> please contact the sender immediately and destroy all copies of 
>>>>>>> the original message.
>>>>>>>
>>>>>
>>>>> -- 
>>>>> *Greg Kibilko*
>>>>> *SETS - SE Technical Sales, Inc.*
>>>>> 112 Tyra Lane
>>>>> Townville, SC 29689
>>>>> P: (864) 356-2606
>>>>> CAEN Products 
>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=GtuKScKnnvQz5mqUxXps-eQDnWdyIH3rVmFwdtCRzW8&s=S7h3K8fgiI00eZJ2kEf5yMIeKJgvTkyxBfOWhilcrqc&e=>
>>>>> CAEN SyS Products & Solutions 
>>>>> <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=GtuKScKnnvQz5mqUxXps-eQDnWdyIH3rVmFwdtCRzW8&s=ejkMDj6QjXOr7xaA_U4V8SBEF3Ty-1iA6DRqqHJIPmU&e=>
>>>>> IOxOs Products 
>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=GtuKScKnnvQz5mqUxXps-eQDnWdyIH3rVmFwdtCRzW8&s=SK--V8r9rxg_h2_07vjN3jo9DU_UFOEVmwAT7ndFXdU&e=>
>>>>
>>>> -- 
>>>> *Greg Kibilko*
>>>> *SETS - SE Technical Sales, Inc.*
>>>> 112 Tyra Lane
>>>> Townville, SC 29689
>>>> P: (864) 356-2606
>>>> CAEN Products 
>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=8n85U3kspPySw_6XgdPf_7vG1P-esCOYDMN-HlhRyTc&s=lb66aikn92Ry8Wk9SaQ8vFQizV9CzQKqezE8YwZi0gg&e=>
>>>> CAEN SyS Products & Solutions 
>>>> <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=8n85U3kspPySw_6XgdPf_7vG1P-esCOYDMN-HlhRyTc&s=C0T3o7YcsYSeDSiX-EqAjmW1uJ_jYH_AlvzkicfbySA&e=>
>>>> IOxOs Products 
>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=8n85U3kspPySw_6XgdPf_7vG1P-esCOYDMN-HlhRyTc&s=NGpHRCiyokCFKbLwHfQR3HDyRk33sGD8AmL6QmU1MZQ&e=>
>>>
>>> -- 
>>> *Greg Kibilko*
>>> *SETS - SE Technical Sales, Inc.*
>>> 112 Tyra Lane
>>> Townville, SC 29689
>>> P: (864) 356-2606
>>> CAEN Products 
>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=vbwUFscQkc7KTGDe9GAjeX-t3VQrEAl1XViPCqzSzV8&s=4xfSXkN5Z0KOWScTwuPvv-GLuFtmovQUoxA2at8uBKE&e=>
>>> CAEN SyS Products & Solutions 
>>> <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=vbwUFscQkc7KTGDe9GAjeX-t3VQrEAl1XViPCqzSzV8&s=gxgoqjk0K9XVim_0nEeqMcYrEZE1XC2myd5_vz1CqrI&e=>
>>> IOxOs Products 
>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=vbwUFscQkc7KTGDe9GAjeX-t3VQrEAl1XViPCqzSzV8&s=-8HbCmy7_bGKGU9VkuQc0b2YhPLKdEVVR95wylBj1N4&e=>
>>
>>
>> -- 
>>
>>  CAEN S.p.A.  Gianni Di Maio
>>
>> *Maintenance Division Manager*
>>
>> Email: g.dimaio at caen.it <mailto:g.dimaio at caen.it>
>>
>> Skype: gdimaio_caen
>>
>> 	Cell:+393482556525 	
>>
>> *CAEN S.p.A.*
>>
>> Tel.: +39 0584 388 398
>> Fax: +39 0584 388 959
>> Email: info at caen.it <mailto:info at caen.it>
>> via Vetraia,11 • 55049 • Viareggio (LU) • Italy
>> Cod.Fisc. e p.IVA IT00864500467
>> https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwIF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=Y_6u1GUQugVDYN3xLTG55vJvalWvRQDGwLbMtQADg4g&s=v2iX4fTA7dIEEaBAAVjieKWdSnGj_t_6h3WZCaNfS-0&e=  
>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=vbwUFscQkc7KTGDe9GAjeX-t3VQrEAl1XViPCqzSzV8&s=o9Ni7h-V49aiCSdssSSjdKGUA-UYCkHGR3IDa7KGKq0&e=> 
>>
>>
>> Le informazioni trasmesse attraverso la presente comunicazione via 
>> mail ed eventuali allegati sono da intendersi di esclusiva spettanza 
>> del destinatario. Nel caso in cui le stesse raggiungessero, per 
>> qualunque motivo, soggetti diversi dai destinatari, questi dovrebbero 
>> darne immediata notizia al mittente. In ogni caso, eventuali soggetti 
>> diversi dai legittimi destinatari della presente comunicazione sono 
>> esplicitamente diffidati da ogni utilizzazione, sia ai sensi del 
>> D.Lgs. n. 196/2003 "Codice in materia di protezione dei dati 
>> personali" sia ai sensi degli art. 616 e ss. Codice Penale che 
>> sanziona la violazione del segreto sulla corrispondenza.
>> The information contained in this transmission may contain privileged 
>> and confidential information and is intended only for the use of the 
>> person(s)named above. If you are not the intended recipient, or an 
>> employee or agent responsible for delivering this message to the 
>> intended recipient, any review, dissemination, distribution or 
>> duplication of this communication is strictly prohibited. If you are 
>> not the intended recipient, please contact the sender immediately and 
>> destroy all copies of the original message.
>>
>
> -- 
>
>  CAEN S.p.A.  Gianni Di Maio
>
> *Maintenance Division Manager*
>
> Email: g.dimaio at caen.it <mailto:g.dimaio at caen.it>
>
> Skype: gdimaio_caen
>
> 	Cell:+393482556525 	
>
> *CAEN S.p.A.*
>
> Tel.: +39 0584 388 398
> Fax: +39 0584 388 959
> Email: info at caen.it <mailto:info at caen.it>
> via Vetraia,11 • 55049 • Viareggio (LU) • Italy
> Cod.Fisc. e p.IVA IT00864500467
> https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwIF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=Y_6u1GUQugVDYN3xLTG55vJvalWvRQDGwLbMtQADg4g&s=v2iX4fTA7dIEEaBAAVjieKWdSnGj_t_6h3WZCaNfS-0&e=  
> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=l1Bxe6M2FwPRKOHa8HakJJbq1qyFdUuuRxh7H5fAcFE&s=EaMVA55syYs25lBL7K8td4XJi1yiNirDJ2mvVT8lzd4&e=> 
>
>
> Le informazioni trasmesse attraverso la presente comunicazione via 
> mail ed eventuali allegati sono da intendersi di esclusiva spettanza 
> del destinatario. Nel caso in cui le stesse raggiungessero, per 
> qualunque motivo, soggetti diversi dai destinatari, questi dovrebbero 
> darne immediata notizia al mittente. In ogni caso, eventuali soggetti 
> diversi dai legittimi destinatari della presente comunicazione sono 
> esplicitamente diffidati da ogni utilizzazione, sia ai sensi del 
> D.Lgs. n. 196/2003 "Codice in materia di protezione dei dati 
> personali" sia ai sensi degli art. 616 e ss. Codice Penale che 
> sanziona la violazione del segreto sulla corrispondenza.
> The information contained in this transmission may contain privileged 
> and confidential information and is intended only for the use of the 
> person(s)named above. If you are not the intended recipient, or an 
> employee or agent responsible for delivering this message to the 
> intended recipient, any review, dissemination, distribution or 
> duplication of this communication is strictly prohibited. If you are 
> not the intended recipient, please contact the sender immediately and 
> destroy all copies of the original message.
>

-- 

  CAEN S.p.A.  Gianni Di Maio

*Maintenance Division Manager*

Email: g.dimaio at caen.it

Skype: gdimaio_caen

	Cell:+393482556525 	

*CAEN S.p.A.*

Tel.: +39 0584 388 398
Fax: +39 0584 388 959
Email: info at caen.it
via Vetraia,11 • 55049 • Viareggio (LU) • Italy
Cod.Fisc. e p.IVA IT00864500467
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwIF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=Y_6u1GUQugVDYN3xLTG55vJvalWvRQDGwLbMtQADg4g&s=v2iX4fTA7dIEEaBAAVjieKWdSnGj_t_6h3WZCaNfS-0&e= 

Le informazioni trasmesse attraverso la presente comunicazione via mail 
ed eventuali allegati sono da intendersi di esclusiva spettanza del 
destinatario. Nel caso in cui le stesse raggiungessero, per qualunque 
motivo, soggetti diversi dai destinatari, questi dovrebbero darne 
immediata notizia al mittente. In ogni caso, eventuali soggetti diversi 
dai legittimi destinatari della presente comunicazione sono 
esplicitamente diffidati da ogni utilizzazione, sia ai sensi del D.Lgs. 
n. 196/2003 "Codice in materia di protezione dei dati personali" sia ai 
sensi degli art. 616 e ss. Codice Penale che sanziona la violazione del 
segreto sulla corrispondenza.
The information contained in this transmission may contain privileged 
and confidential information and is intended only for the use of the 
person(s)named above. If you are not the intended recipient, or an 
employee or agent responsible for delivering this message to the 
intended recipient, any review, dissemination, distribution or 
duplication of this communication is strictly prohibited. If you are not 
the intended recipient, please contact the sender immediately and 
destroy all copies of the original message.

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallc_controls/attachments/20200302/49e36b89/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: agopfchfhhhjpfcp.png
Type: image/png
Size: 554509 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallc_controls/attachments/20200302/49e36b89/attachment-0001.png>


More information about the Dsg-hallc_controls mailing list