[dsg-hdice] Status
Dr. A.M. Sandorfi
sandorfi at jlab.org
Thu Jan 26 18:17:21 EST 2017
Dear Amanda,
The attached file contains three sections. (We've highlighted a few specific
places where we would appreciate feedback.)
(I) a summary of Xiangdong's main points from Dec 5/16. As we send such
notes back and forth, it would be helpful to "check off" the points when
completed, and note the extent of the testing.
(II) the required range of the user-defined parameters that are loaded into
the setup table for the NMRS program.
(III) a discussion of our synchronization requirements, with questions as to
what options you are pursuing.
Thank you,
Andy Sandorfi and Xiangdong Wei
On 1/20/17, 11:33 AM, "Amanda Hoebel" <amandah at jlab.org> wrote:
> Dear Dr. Sandorfi,
Thank you for your email. With regards to the 3 items
> discussed in the meeting (functionality of Rack 2, pump cart diagram, and
> synchronization of the lock-in amplifier data), we are awaiting the detailed
> list for the work to be done.
The following have been completed for Rack 2
> functionality:
1. Updated FRS and NMR programs to LabVIEW 2015.
2. Added
> CT-box to NMR program.
3. Replaced old cables with new low-noise cables.
> 4. Incorporated LCD screen on RF Attenuation box.
5. Modified NMR program
> to run scans with positive and negative current.
6. Added Tbottom and Twait
> times to NMR program.
7. Made Tup and Tdown independent of each other (for
> 10 s 300 s).
We need to know what else needs to be focused on in order to
> make Rack 2 completely functional.
The power-portion of the pump-cart wiring
> diagram has been completed. The signal-portion of the diagram has not been
> completed. We believe that the diagram you provided us is more of a block
> diagram than a wiring diagram; however, we have removed the locks, replaced
> the UPS batteries, and reconnected the power cords to the pump cart. We¹ve
> requested Mike Lowry to check the connections.
For the synchronization of
> the lock-in amplifier, details are needed for the trigger and acquisition
> rate. We need the limits for all parameters (maximum and minimum times needed
> for Tdown, Tbottom, Tup, and Twait), as the program can only acquire data at a
> fixed rate calculated by the total time requested and buffer depth of the
> lock-in amplifier.
Further, we believe the NMR program should be thoroughly
> tested before the synchronization program can begin, to establish a baseline
> for how the program should function. At present, we estimate that the
> synchronization program should take us about 4 months, if we are provided with
> the necessary tools.
Please provide us with this information and we will
> work to have these items
> completed.
Regards,
Amanda
_______________________________________________
d
> sg-hdice mailing
> list
dsg-hdice at jlab.org
https://mailman.jlab.org/mailman/listinfo/dsg-hdice
-------------- next part --------------
A non-text attachment was scrubbed...
Name: NMR parameters -Jan26'17.pdf
Type: application/octet-stream
Size: 866245 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hdice/attachments/20170126/27a4b326/attachment-0002.obj>
More information about the dsg-hdice
mailing list