[Hallb-engineering] [New Logentry] Follow-up Re: Solenoid Magnet Reservoir Pressure Control
kashy at jlab.org
kashy at jlab.org
Mon Dec 18 15:00:02 EST 2017
Logentry Text:
--
The EV8670BY valve had a minimum of 25% from the EV8670BY_min loop. This was changed a few days ago to 10% by Denny to keep the shield from warming too much.
I changed the Input Parameter of the PID to the warmer of the shield outlets from the Magnet (TR8672) and made the set value 43K. I also changed the DP setpoint from 0.2 atm to 0.03 so the valve reacts earlier and quicker to a fast dump and the max change from 10 to 30.
The temperature of TR8672 overshot a bit but onle 1.7K so we will leave this overnight and probably increase it in the morning.
The goal is this is to keep the relief valve from opening during a fast dump of the Solenoid.
---
This is a plain text email for clients that cannot display HTML. The full logentry can be found online at https://logbooks.jlab.org/entry/3506530
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/hallb-engineering/attachments/20171218/726ae007/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: clonpc20_20171218_145307.gif
Type: image/gif
Size: 166110 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/hallb-engineering/attachments/20171218/726ae007/attachment-0002.gif>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: clonpc20_20171218_145452.gif
Type: image/gif
Size: 74616 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/hallb-engineering/attachments/20171218/726ae007/attachment-0003.gif>
More information about the Hallb-engineering
mailing list