This summer we moved the Superconducting liquid helium probe higher in the TST Lead reservoir. Today I tested the system to see if the issue with backfilling of the recoolers still occurs and at what liquid levels on both level probes LL8210SC and LL8120DP.

Conclusions:
1)  Batch filling still occurs if the thermosiphon return stand pipe in the TST is not covered with LHe. 
2) The SC probe min value before batchfilling occurs 63% and the DP probe is 57%. 
3) We updated the controlled rampdown values for the Torus to <70% on the SC probe and <65% on the DP probe, (note both were 40% which was too low). 
4) We will operate the Torus at 90% Liquid level on the SC probe which is about 70% on the DP probe (the level we operated on last spring).



---

This is a plain text email for clients that cannot display HTML.  The full logentry can be found online at https://logbooks.jlab.org/entry/3590235
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/hallb-engineering/attachments/20180822/f0cc34cd/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: clonpc14_20180822_161959.gif
Type: image/gif
Size: 42307 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/hallb-engineering/attachments/20180822/f0cc34cd/attachment-0002.gif>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: clonpc14_20180822_162429.gif
Type: image/gif
Size: 123373 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/hallb-engineering/attachments/20180822/f0cc34cd/attachment-0003.gif>


More information about the Hallb-engineering mailing list