[Halld-cal] FCAL CW bases
Fernando J. Barbosa
barbosa at jlab.org
Fri Jun 18 11:24:28 EDT 2010
Hi Matt,
I have a few concerns about the plan for the CW bases you suggested at
the GlueX meeting and regarding the HV sparks observed on the present
design:
1. HV sparks across components - Due to close proximity of components,
HV sparks or leakage are observed. The present proposed solution is to
pot with epoxy. This seems to prevent HV sparks.
My concern is that this is temporary solution but not a cure for the
inherent problem. I really don't want to rely on potting with epoxy
(long term issues) and a conformal coating should only be used as a
preventive measure.
2. HV sparks are observed across one HV capacitor - the capacitor is
rated for operation at 2,500V but it cannot hold 1,500V. The proposed
solution is to use a low viscosity epoxy to under fill the capacitor and
prevent sparking.
A capacitor that is sold and rated at 2,500V should withstand that same
voltage or the manufacturer has a big problem. Most likely the PCB pads
to which the cap is soldered to are too close or have sharp edges - this
limits the maximum operating voltage. Alternatively, you could choose a
larger capacitor with the pads spaced farther apart but this may not be
necessary.
I strongly suggest that the base be re-designed to space the components
and pads farther apart to provide for reliable operation by design,
without a conformal coating. Once that is verified, a conformal coating
is highly recommended because of moisture and dust.
Best regards,
Fernando