[Halld-cal] [EXTERNAL] BCAL calibration status

Zisis Papandreou zisis at uregina.ca
Wed Apr 1 10:48:38 EDT 2020


You got it.  Thanks and cheers, Zisis…

> On Apr 1, 2020, at 5:34 AM, Elton Smith <elton at jlab.org> wrote:
> 
> Hi Zisis and Karthik,
> 
> Thanks for the effort to complete the gain calibrations and keeping us up-to-date. I am replying to the halld-cal list, which is where future communications should be sent so that everyone is informed and also we keep a record of the communications.
> 
> Cheers, Elton
> 
> Elton Smith
> Jefferson Lab MS 12H3
> 12000 Jefferson Ave STE 4
> Newport News, VA 23606
> (757)269-7625
> (757)269-6331 fax
> 
> On 3/31/20 8:27 PM, Zisis Papandreou wrote:
>> Hi Sean:
>> 
>> as you saw from Karthik’s email, he started with the fall 2018 gains, then applied Mark’s ratio, and started iterations.  He is looking at Batch 1, runs 71350-71591. (As this batch was for 150/250 nA, we will look at the other batches at 350 nA later, to see if any tweaks in the gains will be needed.)
>> 
>> 6 iterations have completed.  The file is attached below (to make it smaller, Karthik removed the 2D plots of 2g vs channel no; those are available upon request.)
>> 
>> It looks like it takes ~3 hrs to do an iteration, so by tomorrow we should have 2-3 more iterations, and then see if the top left plot (2g vs channel no fits) converges to a line, for all three layers.
>> 
>> Cheers, Zisis...
>> 
>> 
>>> On Mar 31, 2020, at 3:00 PM, Sean Dobbs <sdobbs at fsu.edu <mailto:sdobbs at fsu.edu>> wrote:
>>> 
>>> Great news, thanks!
>>> I suppose this is starting with flat gains?
>>> 
>>> Cheers,
>>> Sean
>>> 
>>> 
>>> On Tue, Mar 31, 2020 at 4:58 PM Zisis Papandreou <zisis at uregina.ca <mailto:zisis at uregina.ca>> wrote:
>>> 
>>>    Hi Sean:
>>> 
>>>    Karthik reports: /The calibration for spring 2020 (batch 1) is
>>>    going smooth and I have completed 4 iterations for now and 5th
>>>    iteration is running in the farm. I am attaching the plots that
>>>    we usually produce when we do the calibration./
>>> 
>>>    The file was very large and Karthik will modify it to make it
>>>    easier to work with. Just to show you that things are appearing
>>>    to converge, I am attaching a couple of screen shots.
>>> 
>>>    Karthik is now estimating how lone it will take to finish the 5th
>>>    iteration, and we will see whether we need to go to 8, or whether
>>>    we can live with fewer.
>>> 
>>>    We will continue to keep you posted.
>>> 
>>>    Cheers, Zisis...
>>> 
>>>    <mean and sigma.png>
>>> 
>>>    <gain constants.png>
>>> 
>>>>    On Mar 30, 2020, at 10:55 AM, Sean Dobbs <sdobbs at fsu.edu
>>>>    <mailto:sdobbs at fsu.edu>> wrote:
>>>> 
>>>>    Hi Zisis,
>>>> 
>>>>    Thanks very much for the update.  I prefer to have the BCAL gains at
>>>>    an advanced stage, if not complete, before we do the next monitoring
>>>>    launch.  So let's plan to keep this a high priority and hope to
>>>>    launch
>>>>    on Friday or Monday if need be.
>>>> 
>>>>    Cheers,
>>>>    Sean
>>>> 
>>>>    On Mon, Mar 30, 2020 at 12:50 PM Zisis Papandreou
>>>>    <zisis at uregina.ca <mailto:zisis at uregina.ca>> wrote:
>>>>> 
>>>>>    Hi Sean:
>>>>> 
>>>>>    we had a another Regina group meeting this morning, where
>>>>>    Karthik reported the activities.  Tegan also joins these.
>>>>> 
>>>>>    1) PrimEx jobs: jobs are finally running on the farm.  Igal
>>>>>    gave Karthik some advise on how to tune job requirements so
>>>>>    that the jobs can run faster on the farm.
>>>>> 
>>>>>    2) GlueX jobs:
>>>>>    - Two test jobs completed without any issues.  The mass is
>>>>>    coming out from the fits at 135 MeV/c2.  There are not enough
>>>>>    statistics to see things cell by cell or to compare the layers,
>>>>>    but it looks good so far.
>>>>>    - The fall 2019 pi0 calibration took 6 iterations to converge,
>>>>>    which were completed in less than a week while the farm was not
>>>>>    too busy.  Tegan noted that whether one starts from flat gains
>>>>>    or previous ccdb gains, the convergence of the iterations takes
>>>>>    about the same amount of time.
>>>>> 
>>>>>    Question: Sean et al: it is not likely that the GlueX
>>>>>    jobs/iterations will complete by April 1.  We hope that more
>>>>>    jobs will complete by tomorrow, perhaps offering us enough
>>>>>    stats to see the layers fits and cells.  If this happens *and*
>>>>>    the results look good, we could discuss whether to postpone the
>>>>>    launch towards the end of the week, when we hope the
>>>>>    calibration should be close to complete. Thoughts?
>>>>> 
>>>>>    Cheers, Zisis...
>>>>> 
>>>>>    On Mar 27, 2020, at 1:29 PM, Sean Dobbs <sdobbs at fsu.edu
>>>>>    <mailto:sdobbs at fsu.edu>> wrote:
>>>>> 
>>>>>    Hi Zisis and Karthik,
>>>>> 
>>>>>    Thanks a lot for the report.  We will keep an eye on how things
>>>>>    progress and modify the schedule accordingly.  My jobs have been
>>>>>    sitting there forever as well.
>>>>> 
>>>>>    Maybe we should discuss bumping up Karthik's priority on the farm,
>>>>>    since this is urgent?
>>>>>    An additional idea is that after this iteration, we can pull
>>>>>    the files
>>>>>    from tape, pin them to the cache disk, and then he can be sure that
>>>>>    they are all on disk for future iterations.
>>>>> 
>>>>>    Cheers,
>>>>>    Sean
>>>>> 
>>>>>    On Fri, Mar 27, 2020 at 2:32 PM Zisis Papandreou
>>>>>    <zisis at uregina.ca <mailto:zisis at uregina.ca>> wrote:
>>>>> 
>>>>> 
>>>>>    Hi everyone:
>>>>> 
>>>>>    we had a group meeting this morning in Regina to discuss the
>>>>>    PrimEx and GlueX BCAL calibrations.
>>>>> 
>>>>>    PrimEx: Karthik did a quick test with runs from He with CDC on
>>>>>    and sees good pi0 peaks in the Layer 1 and 2 cells, but not in
>>>>>    3 and barely anything in 4 (the last one is not expected, due
>>>>>    to low penetration of showers there). The calibration from past
>>>>>    GlueX runs does not show this L3 behaviour. We are still
>>>>>    investigating, including checking how well the skims work, when
>>>>>    there is no B-field on.  Also, Karthik will look at hdview to
>>>>>    see if anything interesting is observable.  Assuming we fix L3,
>>>>>    then calibration can proceed, where we will have to bump up the
>>>>>    pi0 mass from 130 to 135.  This will be a PrimEx-specific set
>>>>>    of gains, entered in a suitable place in ccdb.
>>>>> 
>>>>>    GlueX: Sean, see Karthik’s email below.  H checked and his
>>>>>    account is ok, no holds on it, proper use of sqlite, etc, after
>>>>>    getting advice from Mark Ito on how to do all that. As the jobs
>>>>>    are at a virtual standstill, it does not look like we can make
>>>>>    the April 1 launch.  Of course, we can wait over the weekend
>>>>>    and see if by Monday some of the jobs get executed, and then
>>>>>    decide what to do then (go ahead, or delay the launch).
>>>>> 
>>>>>    Comments, thoughts?
>>>>> 
>>>>>    Cheers, Zisis…
>>>>> 
>>>>>    =============
>>>>> 
>>>>> 
>>>>>    Good morning,
>>>>> 
>>>>>    I applied the gainratios to the existing ccdb gains and submitted a
>>>>>    small test batch on the 25th of march (wednesday). The jobs has
>>>>>    been in
>>>>>    the depend state since then. I understand that this is because
>>>>>    of a lot
>>>>>    of jobs being submitted to the farm lately. However, if this
>>>>>    continues,
>>>>>    I am not sure if we can deliver the gains by April 1. I shall
>>>>>    continue
>>>>>    to do the gains and deliver it at the latest. This is in the
>>>>>    top of my
>>>>>    priority list.
>>>>> 
>>>>>    Thank you
>>>>> 
>>>>>    Karthik Suresh
>>>>> 
>>>>>    On Mar 25, 2020, at 12:20 PM, Igal Jaegle <ijaegle at jlab.org
>>>>>    <mailto:ijaegle at jlab.org>> wrote:
>>>>> 
>>>>>    Thank you Zizis & Karthik, your help is much appreciated.
>>>>> 
>>>>>    tks ig.
>>>>>    ________________________________
>>>>>    From: Zisis Papandreou <Zisis.Papandreou at uregina.ca
>>>>>    <mailto:Zisis.Papandreou at uregina.ca>>
>>>>>    Sent: Wednesday, March 25, 2020 1:29 PM
>>>>>    To: sdobbs at fsu.edu <mailto:sdobbs at fsu.edu> <sdobbs at fsu.edu
>>>>>    <mailto:sdobbs at fsu.edu>>; Mark-Macrae Dalton <dalton at jlab.org
>>>>>    <mailto:dalton at jlab.org>>; Elton Smith <elton at jlab.org
>>>>>    <mailto:elton at jlab.org>>; gasparan at jlab.org
>>>>>    <mailto:gasparan at jlab.org> <gasparan at jlab.org
>>>>>    <mailto:gasparan at jlab.org>>; Igal Jaegle <ijaegle at jlab.org
>>>>>    <mailto:ijaegle at jlab.org>>; Ilya Larin <ilarin at jlab.org
>>>>>    <mailto:ilarin at jlab.org>>; Jonathan Zarling <jzarling at jlab.org
>>>>>    <mailto:jzarling at jlab.org>>; lgan at jlab.org
>>>>>    <mailto:lgan at jlab.org> <lgan at jlab.org <mailto:lgan at jlab.org>>;
>>>>>    Alexander Somov <somov at jlab.org <mailto:somov at jlab.org>>;
>>>>>    ksv656 at uregina.ca <mailto:ksv656 at uregina.ca><ksv656 at uregina.ca
>>>>>    <mailto:ksv656 at uregina.ca>>
>>>>>    Subject: Re: [EXTERNAL] Re: BCAL calibration status
>>>>> 
>>>>>    Hi Sasha and Igal:
>>>>> 
>>>>>    thanks for summarizing the PrimEx goals.  This is helpful.
>>>>> 
>>>>>    Sorry that these calibrations had not bee done earlier, but
>>>>>    much like City Officials, we are "complaint" driven. I wish we
>>>>>    were more proactive.
>>>>> 
>>>>>    Anyhow, rest assured that this has a high priority now.     Karthik is juggling PrimEx and the GlueX calibration (to meet
>>>>>    the launch date).   On the former, he may have some plots later
>>>>>    in the day, and if these look ok, he can turn the crank and do
>>>>>    all the runs.
>>>>> 
>>>>>    We will keep you posted.
>>>>> 
>>>>>    Cheers, Zisis...
>>>>> 
>>>>> 
>>>>> 
>>>>>    Dr. Zisis Papandreou
>>>>>    Head, Department of Physics
>>>>>    Professor of Physics, Ph.D, P.Phys.
>>>>>    ---------------------------------------------------------
>>>>>    Department of Physics
>>>>>    University of Regina
>>>>>    3737 Wascana Parkway
>>>>>    Regina, SK  S4S 0A2     CANADA
>>>>> 
>>>>>    Phone: (306) 585-5379
>>>>>    Fax: (306) 585-5659
>>>>>    Email: zisis at uregina.ca <mailto:zisis at uregina.ca>
>>>>>    Website:
>>>>>    https://urldefense.proofpoint.com/v2/url?u=https-3A__www.uregina.ca_science_physics_people_faculty-2Dresearch_zisis-2Dpapandreou_&d=DwIFaQ&c=CJqEzB1piLOyyvZjb8YUQw&r=qOykDDW07dxNIz2gqzRW_0sRJeff1E30DnojDqga1Jo&m=IraULg_SMVRy6o10CAjzyiksMbHSlSahZEvO70xul8U&s=y0CSWvLVbEiZEk_XK0U5rCnWl77kf4eeCdZctW_bbDk&e= 
>>>>>    <https://urldefense.com/v3/__https://www.uregina.ca/science/physics/people/faculty-research/zisis-papandreou/__;%21%21PhOWcWs%21lYKDfQmd6uehc2l57agVvudr58rCki-gmq37OV_v-mjsS3gAi8qT_jR1mLeUAA$>
>>>>> 
>>>>>    Alexander Somov <somov at jlab.org <mailto:somov at jlab.org>>
>>>>>    03/17/20 8:05 PM >>>
>>>>> 
>>>>>    Hi Zisis, All,
>>>>> 
>>>>>    I apologize for the late response, too many things
>>>>>    today.
>>>>> 
>>>>>    As we have discussed, the BCAL calibration precision
>>>>>    for PrimEx runs should be the same as for other GlueX
>>>>>    runs. At least there should be no shifts in mass
>>>>>    distributions of pi0, etc. We don't request anything
>>>>>    special.
>>>>> 
>>>>>    BCAL will be used to subtract background from
>>>>>    hadronic interactions.
>>>>> 
>>>>>    BCAL is a part of the PrimEx equipment, and PrimEx
>>>>>    is the GlueX experiment, so we expect to
>>>>>    receive support from the collaboration
>>>>>    (BCAL group).
>>>>> 
>>>>>    Timeline. I Was pretty much sure that the calibration
>>>>>    has already been done. Elton reported this
>>>>>    when I asked about the BCAL calibration status for
>>>>>    PrimEx runs during the GlueX meeting.
>>>>>    Obviously it was a wrong statement (based on Igal's
>>>>>    checks).
>>>>> 
>>>>>    ----
>>>>> 
>>>>>    I don't know how much work is needed to do calibration. Perhaps
>>>>>    applying some simple
>>>>>    corrections would be enough.
>>>>> 
>>>>>    Cheers,
>>>>>               Sasha
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>>    ________________________________
>>>>>    From: Zisis Papandreou <Zisis.Papandreou at uregina.ca
>>>>>    <mailto:Zisis.Papandreou at uregina.ca>>
>>>>>    Sent: Tuesday, March 17, 2020 9:25 AM
>>>>>    To: sdobbs at fsu.edu <mailto:sdobbs at fsu.edu> <sdobbs at fsu.edu
>>>>>    <mailto:sdobbs at fsu.edu>>; Mark-Macrae Dalton <dalton at jlab.org
>>>>>    <mailto:dalton at jlab.org>>; Elton Smith <elton at jlab.org
>>>>>    <mailto:elton at jlab.org>>; Igal Jaegle <ijaegle at jlab.org
>>>>>    <mailto:ijaegle at jlab.org>>; Jonathan Zarling <jzarling at jlab.org
>>>>>    <mailto:jzarling at jlab.org>>; Alexander Somov <somov at jlab.org
>>>>>    <mailto:somov at jlab.org>>; ksv656 at uregina.ca
>>>>>    <mailto:ksv656 at uregina.ca><ksv656 at uregina.ca
>>>>>    <mailto:ksv656 at uregina.ca>>
>>>>>    Subject: [EXTERNAL] Re: BCAL calibration status
>>>>> 
>>>>>    Hello Igal et al:
>>>>> 
>>>>>    yes, work is being done: jobs are running on the calibration
>>>>>    and MC studies are also being carried out. While we wait for
>>>>>    those, this is a good time to get "organized".
>>>>> 
>>>>>    In our group, we are not sure of what the objectives and
>>>>>    priorities are for the Primex runs, particularly as they
>>>>>    pertain to the BCAL
>>>>> 
>>>>>    Igal, can you or Sasha, perhaps make a few slides or write a
>>>>>    one-pager, explaining those objectives?  It would be good if
>>>>>    there were timelines (of what you are trying to extract from
>>>>>    the data) or the tasks listed in order of priority.  We could
>>>>>    then go over those at this Thursday's Calorimetry meeting.
>>>>> 
>>>>>    The reason we ask for this is that the Regina crew has other
>>>>>    tasks on their plates and we should optimize our collective
>>>>>    operations.  In parallel, Karthik and Varun are studying to
>>>>>    take their Ph.D. Comprehensive exams in May.
>>>>> 
>>>>>    Thanks and best regards, Zisis...
>>>>> 
>>>>> 
>>>>> 
>>>>>    Dr. Zisis Papandreou
>>>>>    Head, Department of Physics
>>>>>    Professor of Physics, Ph.D, P.Phys.
>>>>>    ---------------------------------------------------------
>>>>>    Department of Physics
>>>>>    University of Regina
>>>>>    3737 Wascana Parkway
>>>>>    Regina, SK  S4S 0A2     CANADA
>>>>> 
>>>>>    Phone: (306) 585-5379
>>>>>    Fax: (306) 585-5659
>>>>>    Email: zisis at uregina.ca <mailto:zisis at uregina.ca>
>>>>>    Website:
>>>>>    https://urldefense.proofpoint.com/v2/url?u=https-3A__www.uregina.ca_science_physics_people_faculty-2Dresearch_zisis-2Dpapandreou_&d=DwIFaQ&c=CJqEzB1piLOyyvZjb8YUQw&r=qOykDDW07dxNIz2gqzRW_0sRJeff1E30DnojDqga1Jo&m=IraULg_SMVRy6o10CAjzyiksMbHSlSahZEvO70xul8U&s=y0CSWvLVbEiZEk_XK0U5rCnWl77kf4eeCdZctW_bbDk&e= 
>>>>>    <https://urldefense.com/v3/__https://www.uregina.ca/science/physics/people/faculty-research/zisis-papandreou/__;%21%21PhOWcWs%21lYKDfQmd6uehc2l57agVvudr58rCki-gmq37OV_v-mjsS3gAi8qT_jR1mLeUAA$>
>>>>> 
>>>>>    Igal Jaegle <ijaegle at jlab.org <mailto:ijaegle at jlab.org>>
>>>>>    03/17/20 3:59 AM >>>
>>>>> 
>>>>>    BCAL colleagues,
>>>>> 
>>>>>    Do you have any news for the BCAL calibration of the PrimEX-D
>>>>>    measurements?
>>>>> 
>>>>>    tks ig.
>>>>> 
>>>>> 
>>>>> 
>>>>>    Dr. Zisis Papandreou
>>>>>    Head, Department of Physics
>>>>>    Professor of Physics, Ph.D., P.Phys.
>>>>>    ---------------------------------------------------------
>>>>>    Department of Physics
>>>>>    University of Regina
>>>>>    3737 Wascana Parkway
>>>>>    Regina, SK  S4S 0A2  CANADA
>>>>> 
>>>>>    Phone: (306) 585-5379
>>>>>    Fax: (306) 585-5659
>>>>>    Email: zisis at uregina.ca <mailto:zisis at uregina.ca>
>>>>>    Website:
>>>>>    https://urldefense.proofpoint.com/v2/url?u=http-3A__www.uregina.ca_science_physics_people_faculty-2Dresearch_zisis-2Dpapandreou_index.html&d=DwIFaQ&c=CJqEzB1piLOyyvZjb8YUQw&r=qOykDDW07dxNIz2gqzRW_0sRJeff1E30DnojDqga1Jo&m=IraULg_SMVRy6o10CAjzyiksMbHSlSahZEvO70xul8U&s=pa_RD1qwpbm7Bz3DvnBarc4rkv8Wk-e7WOv4ydBypbE&e= 
>>>>>    <https://urldefense.com/v3/__http://www.uregina.ca/science/physics/people/faculty-research/zisis-papandreou/index.html__;%21%21PhOWcWs%21lYKDfQmd6uehc2l57agVvudr58rCki-gmq37OV_v-mjsS3gAi8qT_jQSb9zAng$>
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>>    Dr. Zisis Papandreou
>>>>>    Head, Department of Physics
>>>>>    Professor of Physics, Ph.D., P.Phys.
>>>>>    ---------------------------------------------------------
>>>>>    Department of Physics
>>>>>    University of Regina
>>>>>    3737 Wascana Parkway
>>>>>    Regina, SK  S4S 0A2  CANADA
>>>>> 
>>>>>    Phone: (306) 585-5379
>>>>>    Fax: (306) 585-5659
>>>>>    Email: zisis at uregina.ca <mailto:zisis at uregina.ca>
>>>>>    Website:
>>>>>    https://urldefense.proofpoint.com/v2/url?u=http-3A__www.uregina.ca_science_physics_people_faculty-2Dresearch_zisis-2Dpapandreou_index.html&d=DwIFaQ&c=CJqEzB1piLOyyvZjb8YUQw&r=qOykDDW07dxNIz2gqzRW_0sRJeff1E30DnojDqga1Jo&m=IraULg_SMVRy6o10CAjzyiksMbHSlSahZEvO70xul8U&s=pa_RD1qwpbm7Bz3DvnBarc4rkv8Wk-e7WOv4ydBypbE&e= 
>>>>>    <https://urldefense.com/v3/__http://www.uregina.ca/science/physics/people/faculty-research/zisis-papandreou/index.html__;%21%21PhOWcWs%21lYKDfQmd6uehc2l57agVvudr58rCki-gmq37OV_v-mjsS3gAi8qT_jQSb9zAng$>
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>>>> 
>>> 
>> 
>> 
>> 
>> 
>> Dr. Zisis Papandreou
>> Head, Department of Physics
>> Professor of Physics, Ph.D., P.Phys.
>> ---------------------------------------------------------
>> Department of Physics
>> University of Regina
>> 3737 Wascana Parkway
>> Regina, SK S4S 0A2 CANADA
>> 
>> *Phone: (306) 585-5379*
>> Fax: (306) 585-5659
>> Email: zisis at uregina.ca <mailto:zisis at uregina.ca>
>> Website: https://urldefense.proofpoint.com/v2/url?u=http-3A__www.uregina.ca_science_physics_people_faculty-2Dresearch_zisis-2Dpapandreou_index.html&d=DwIFaQ&c=CJqEzB1piLOyyvZjb8YUQw&r=qOykDDW07dxNIz2gqzRW_0sRJeff1E30DnojDqga1Jo&m=IraULg_SMVRy6o10CAjzyiksMbHSlSahZEvO70xul8U&s=pa_RD1qwpbm7Bz3DvnBarc4rkv8Wk-e7WOv4ydBypbE&e=  <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.uregina.ca_science_physics_people_faculty-2Dresearch_zisis-2Dpapandreou_index.html&d=DwMFaQ&c=CJqEzB1piLOyyvZjb8YUQw&r=c-_tH8evYLC4MjBfbp0hiQ&m=L9K-OoIwqk5DKjlr1cfFyqg-V-On_gBLT7g_-xpoYVw&s=YY00N7HeNHCOuCtrD_QDLjdQMkkLKhs2Wd-UwIQijKk&e=>
>> 
>> 
>> 
>> 
>> 
>> 
>> 
> 




More information about the Halld-cal mailing list