[Halld-pid] ST First Tests
Fernando J Barbosa
barbosa at jlab.org
Mon Jul 16 22:34:04 EDT 2012
Hi Mark, Werner,
I have been optimizing and testing the ST readout for the 50 um and 100
um SiPMs. There are 4 SiPMs connected together (current sum) to one
readout channel on each of two SiPM PCBs (50 um and 100 um). The sensor
capacitance is the same in both cases so one optimization is all that is
required for fast risetimes for the group of 4 SiPMs.
The prototype is fully functional but I determined that the micro-coax
ribbon cable assembly (connects the readout to the breakout PCB) is not
good enough due to cross-talk (primarily due to the connectors on these
cable assemblies - a single micro-coax cable assembly provides the power
to the electronics, 3 bias supplies, 6 outputs [3 channels w/ 3 for ADCs
and 3 for TDCs] and has a thermocouple two-wire output for readout). The
readout is temperature compensated via thermistors. The workaround has
two coax cables (ADC and TDC), in addition to the micro-coax, which
bypass the micro-coax signal connections. So, the final design will
change but the current prototype will be useful for tests with ST
detector prototype.
I have attached two pics: ST50 is for the 50 um and ST100 is for the 100
um. These were taken with a laser and a diffuser stimulating all 4 SiPMs
in the group. Some comments:
1. 50 um - Fast pulses with width of 37 ns (~10%-10%)
2. 100 um - Fast pulses with width of 125 ns (~10%-10%), step/kink ~ 5
ns after peak.
The leading edge slew rate is about the same for both but the 100 um has
a slower recovery w/a kink (the quench resistor on each pixel would need
to be ~3x smaller for similar recovery but there are other factors at
play, i.e. Geiger current) . I will investigate this a little further
but these are definitely characteristics of these SiPMs and I will write
a GlueX note soon.
In the meantime, let me know if there is interest in pursuing the 100 um
further. Even though the rates are expected to be in the KHz range, the
long recovery time will likely lead to pileup.
Best regards,
Fernando
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ST50.png
Type: image/png
Size: 85427 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/halld-pid/attachments/20120716/be0f00eb/attachment-0004.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: sT100.png
Type: image/png
Size: 89972 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/halld-pid/attachments/20120716/be0f00eb/attachment-0005.png>
More information about the Halld-pid
mailing list