[Halld-pid] Question on FADC250 clipping
Alexander Somov
somov at jlab.org
Mon Jun 24 15:47:22 EDT 2013
Hi Sascha,
To set the baseline (pedestal) you can use
faSetDAC(ADC_SLOT,3200,0) if you are using coda
(usually the value 3200 brings the pedestal down to the ADC cnt 100 - 300,
just play with values).
Or just write to the register if you are using labview or smth else
See register description at
https://coda.jlab.org/wiki/Downloads/docs/manuals//Programming%20the%20FADCV2_3.pdf
DAC 1..16 /* 0x50 */
Note:
-----
The flash is DC coupled, it can override your discriminator thresholds if
you have a passive splitter (though for a 2 V range the effect should be
small).
Good luck,
Sascha
On Mon, 24 Jun 2013, Elton Smith wrote:
> Hi Sasha,
>
> Just a note to note that the ADC max is 4095 (the 8192 is the overflow
> bit), so a pedestal at 2000 is actually half the dynamic range. We will
> need to adjust pedestal values to a small number in order to preserve
> the dynamic range.
>
> Cheers, Elton.
>
> Elton Smith
> Jefferson Lab MS 12H5
> 12000 Jefferson Ave STE 16
> Newport News, VA 23606
> (757) 269-7625
> (757) 269-6331 fax
>
> On 6/24/13 2:54 PM, Beni Zihlmann wrote:
>> Hi Sasha,
>> for each channel there is a jumper that can be set to 0.5 1.0 and 2.0 Volts.
>> it is very obvious. The jumper is right next to the input on the PCB.
>>
>> cheers,
>> Beni
>>
>>
>>> Hi Fernando,
>>>
>>> In our tests of TOF modules, we have switched now to FADC250
>>> which you sent to us. Things are mostly looking good, we see
>>> signal shapes and all ADC controls are working as expected.
>>> However, we see a lot of signal clipping at 8192 (2^13) ceiling.
>>> On the scope, the triggered signals are on the order of a few
>>> hundred mV, with an average of 600 mV. However, more
>>> than 90% of the ADC shapes are clipped on top. As far as
>>> I understand, the ADC max is selectable with some
>>> resistor block to be 0.5V, 1V or 2V. How do I check that
>>> it is not set to 0.5V in our FADC? Is there any way to check
>>> this through VME, or do I have to measure those resistors?
>>> Also, are the pedestals of ~2000 counts (1/4 of the FADC range)
>>> normal and expected for signals without any visible DC offset?
>>>
>>> Sasha
>>> _______________________________________________
>>> Halld-pid mailing list
>>> Halld-pid at jlab.org
>>> https://mailman.jlab.org/mailman/listinfo/halld-pid
>> _______________________________________________
>> Halld-pid mailing list
>> Halld-pid at jlab.org
>> https://mailman.jlab.org/mailman/listinfo/halld-pid
>
> _______________________________________________
> Halld-pid mailing list
> Halld-pid at jlab.org
> https://mailman.jlab.org/mailman/listinfo/halld-pid
>
More information about the Halld-pid
mailing list