[Primexd] [New Logentry] CCAL rates
somov at jlab.org
somov at jlab.org
Thu Dec 6 00:05:02 EST 2018
Logentry Text:
--
We have performed a study of rates in the CCAL detector.
Beam conditions:
Amorphous 10^-4 R.L. radiator, 750 um converter, beam current 100 - 350 nA.
(PrimEx nominal current - 320 nA assuming runs with the 10^-4 R.L. radiator)
Trigger type: FCAL & CCAL, total energy deposition > 2 GeV
SETUP: hd_all.tsg_ccal_fcal
CONFIG: CCAL_FCAL_TAGGER.conf
Fig. 1 CCAL scalers for the run with 350 nA
Fig. 2 CCAL scalers for the run with 350 nA (energy threshold 30 MeV)
The maximum rate of CCAL modules is 240 kHz for 30 MeV energy threshold.
The maximum rates seems to tolerable (acceptable) for the compton calorimenter.
The rate is relatively close to that predicted by Monte Carlo simulation, see Fig 3.
Fig. 3 Simulation of rates in the Compton calorimeter (slide from the experimental readiness
review).
The PS rate for runs with a 350 nA current is 7.3 kHz (for 750 um converter). The beam transmission
(estimated by comparing PS rates) seems to be 'similar' to that for GlueX runs with the amorphous
radiator. Note, the PS acceptance has to be corrected due to the smaller beam energy of 10.3 GeV.
We took data to check the PMT gain stability on rate. An LED pulser operated at 20 Hz was added to the
production trigger
run 52193 - 350 nA current
run 52194 - 200 nA
run 52195 - 100 nA
run 52196 - no beam
---