It’s now unclear whether we will attempt this today.
puckett.physics.uconn.edu
> On Nov 15, 2021, at 1:11 PM, Evaristo Cisbani <evaristo.cisbani at roma1.infn.it> wrote:
>
> Dear Ezekiel,
>
> 20% of current divider increase should be safe for the divider itself.
>
> According to Dave Mack HV vs current analysis, and similar findings from
> Roberto, the first GEM foil becomes the most critical and at 20 uA of beam
> current it riches ~ 500 V. This is likely still safe: during assembly we
> tested all GEM sectors (in dry conditions) up to 550 V for few tens of minutes.
>
> I am less confident about the foil distortion caused by the larger potential
> between foils (from about 750 V to 900 V); many years ago we did some Finite
> Element evaluations but I did not remember the conclusions and I have been
> unable to retrieve them from my documentation.
>
> As suggested by Kondo and Nilanga, when you go at high beam current, please
> monitor the HV divider current; if it is not stable better switch of the
> chamber immediately.
>
> Thanks
>
> Evaristo
>
>> On Mon, 15 Nov 2021 11:31:40 -0500
>> Ezekiel Wertz <ewertz at email.wm.edu> wrote:
>>
>> Maybe that's more of a question for Evaristo. I do not know. I just want to
>> remind everyone that at 4100V (nominal setting for INFN GEM production) we
>> see about base current with no electron beam of 104 uAmp
>>
>> Zeke
>>
>>> On Mon, Nov 15, 2021 at 11:26 AM Ezekiel Wertz <ewertz at email.wm.edu> wrote:
>>>
>>> Hi Nilanga,
>>>
>>> The INFN GEMs will see probably about a 20-30 uAmp increase in HV at at 20
>>> uAmp of beam current. Is that safe for the divider since excess current
>>> would then be like 20% of the overall current load of the divider?
>>>
>>> Zeke
>>>
>>> On Mon, Nov 15, 2021 at 11:17 AM Liyanage, Nilanga K (nl8n) <
>>> nl8n at virginia.edu> wrote:
>>>
>>>> Hi Sean
>>>>
>>>> Thanks for letting us know.
>>>>
>>>> Yes, it would be OK to get 20 uA on LH2. Please make sure to set the
>>>> current limits accordingly.
>>>>
>>>> Later, for GEP-5 we will have to go to much higher currents, of the order
>>>> of 70 uA; so going to 20 uA now would be a good data point for us.
>>>>
>>>> As Kondo has suggested before, while you are getting this current, please
>>>> monitor the GEM current, if it is not stable and keep going up (after the
>>>> initial increase following the beam current change), you my want to turn
>>>> the chambers off
>>>>
>>>> Best
>>>>
>>>> Nilanga
>>>> ------------------------------
>>>> *From:* Sbs_gems <sbs_gems-bounces at jlab.org> on behalf of Sean Jeffas <
>>>> sj9ry at virginia.edu>
>>>> *Sent:* Monday, November 15, 2021 11:01 AM
>>>> *To:* sbs_gems at jlab.org <sbs_gems at jlab.org>
>>>> *Subject:* [Sbs_gems] [EXTERNAL] Chance for GEM study at 20 uA?
>>>>
>>>> Hi Kondo, Nilanga, and Evaristo,
>>>>
>>>> The MCC is investigating some RF beam issues today, and they said that
>>>> they will send 20 uA of beam to hall A while doing their study. The
>>>> question is, can we keep the GEMs on to see the current draws at 20 uA? We
>>>> would try it on LH2, which has already been tested up to 10 uA. Here are
>>>> the numbers for currents at LH2 that we have taken so far:
>>>>
>>>> No beam: https://logbooks.jlab.org/entry/3943588
>>>> 8 uA on LH2: https://logbooks.jlab.org/entry/3943079
>>>> 10 uA on LD2: Sadly no screenshots were taken, so I will quote the EPICS
>>>> values
>>>> UV Layer 0: 812 uA INFN J0 bot: 116
>>>> UV Layer 2: 804 uA INFN J0 mid: 118
>>>> UVa XY 0: 743 uA INFN J0 top: 112
>>>> UVa XY 1: 768 uA INFN J2 bot: 113
>>>> UVa XY 2: 769 INFN J2 mid: 113
>>>> UVa XY 3: 771 INFN J2 top: 0 (off)
>>>>
>>>>
>>>> So if we go up to 20 uA on LH2 I assume we double the excess currents
>>>> found here we would certainly pass the trip limits on all modules on the
>>>> front tracker. This For example UV layer 0 will go up to probably 870 uA.
>>>> Would this be safe? Are you guys ok with this test?
>>>>
>>>> Best,
>>>> Sean
>>>> _______________________________________________
>>>> Sbs_gems mailing list
>>>> Sbs_gems at jlab.org
>>>> https://mailman.jlab.org/mailman/listinfo/sbs_gems
>>>>
>>>
>
> _______________________________________________
> Sbs_gems mailing list
> Sbs_gems at jlab.org
> https://mailman.jlab.org/mailman/listinfo/sbs_gems
More information about the Sbs_gems
mailing list