[Tpe] TPE shielding
David Kashy
kashy at jlab.org
Thu Jan 27 08:37:46 EST 2011
Larry,
If I understand correctly you want to add shielding on to the platform
that supports the pairspectrometer magnet..
Is this correct?
Dave
On 1/26/2011 5:26 PM, Larry Weinstein wrote:
> Dear Dave,
>
> The wall is 10' wide. That is the approximate distance from the short
> ladder at the pair spectrometer to the upstream end of that platform.
> The wall height would then be beam height plus one foot. (I did not
> measure the height so I am not sure whether it would be 4' or 5'.)
>
> I completely agree that we should not rush into this. I think that
> the radiation experience since Thursday indicates that we do not need
> to do this immediately.
>
> - Larry
>
> David Kashy wrote:
>> Hi Larry,
>> Is this wall 10' high or wide?
>> If it is 10 ft wide then I guess we need to start it above the floor
>> level and need to design a beam structure to support it.
>>
>> If it is 10 ft tall we will still need to design a structure to keep
>> it from toppling.
>>
>> This wall will weight about 8000 lbs (similar to HYCAL) not a trivial
>> load and thus we must do it correctly.
>>
>> Thanks
>> Dave
>>
>>
>>
>> On 1/26/2011 4:55 PM, Larry Weinstein wrote:
>>> Dear Folks,
>>>
>>> Recommendation: In order to be very conservative, build a concrete
>>> wall on the beam left side of the collimator box, frascati magnet 1,
>>> and PS magnet. The wall would be about 10' x 5' x 1' and require
>>> 200 blocks. It should be built during the next extended down time
>>> (Tuesday?) so as not to interfere with taking data.
>>>
>>> Details: I spoke to Pavel Degtiarenko, the radiation and shielding
>>> expert. He says that there is very little useful data about the
>>> effect of radiation on electronics because the response of
>>> electronics to radiation is highly variable. He also noted that the
>>> existing BPM amplifiers have probably already suffered some
>>> radiation damage and that this damage is cumulative. He also said
>>> that most of the radiation is probably high energy particles that
>>> can be shielded by a shield wall (if we know the approximate source
>>> location.)
>>>
>>> This means that we probably want to add shielding to reduce the
>>> probability of catastrophic BPM failure.
>>>
>>> RadCon gave us the badge readings today (thanks, Becky!).
>>>
>>> 66016 331 1' from top of Acc IOC rack
>>> 66017 197 2.5' from top of Acc IOC rack, in front of concrete wall
>>> 66018 220 2.5' from bottom of Acc IOC rack, in front of concrete wall
>>> 66019 66 2.5 from top of Acc IOC rack, behind concrete wall
>>> 66020 73 2' from bottom of Acc IOC rack, behind concrete wall
>>> 66023 92 on ioc ptarg
>>>
>>>
>>> 66024 254 in front of new lead wall, 5' from floor
>>> 66025 191 L1-15 on left (upstream) side of IOC
>>> 66026 632 on railing opposite PS
>>>
>>> 66027 230 upstream, beam side of lead wall on ioctorus
>>> 66028 126 downstream, outside of lead wall on ioc torus
>>>
>>> The first column is the badge number, the second is the radiation
>>> dose, and the third is the location.
>>>
>>> A few notes: Badge 66026 with a reading of 632 mr is located where
>>> the radcon photon probe was located before 1/20, on the railing of
>>> the space frame close to the pair spectrometer. The probe was
>>> reading about 50 mr/hr at that location.
>>>
>>> Badge 66019, behind the concrete wall, is located directly on top of
>>> the accelerator BPM amplifiers and right next to the radcon photon
>>> probe which is currently reading 5 mr/hr. This is consistent with
>>> badge 66026.
>>>
>>> badges 66016, 7, and 8 are located on the accelerator BPM amplifier
>>> rack. The highest one (66016) has the highest dose and has much
>>> less of the lead shielding on the lead wall between it and the beam
>>> line. The concrete wall (either 6" or 12") reduces the rate from
>>> about 200 mr to 70 mr (comparing badges 17 and 18 with 19 and 20).
>>> (Note that the badges were placed before the BPE shielding was added.)
>>>
>>> The maximum thickness of the lead wall reduces the dose from 254
>>> (badge 66024) to 92 (badge 66023).
>>>
>>> The radiation level is not much less a little further down-beam on
>>> rack L1-15 (191 mr at badge 25).
>>>
>>> The dose decreases about a factor of two between the upstream beam
>>> side of the lead wall on ioctorus to the downstream outside of the
>>> lead wall.
>>>
>>> Summary: The dose at iocptarg, ioctorus and the accelerator BPM
>>> amplifiers are within a factor of two. The shielding reduces the
>>> dose significantly.
>>>
>>> There is floor space or floor weight capacity to place any more
>>> shielding near the racks. THis means that the only place to add
>>> shielding is near the source. Since the source probably includes
>>> the collimator box, convertor and chicane, we should build a
>>> concrete wall on the beam left side from the PS magnet going as far
>>> upstream as possible.
>>>
>>> --
>>> Sincerely,
>>> Larry
>>>
>>> -----------------------------------------------------------
>>> Lawrence Weinstein
>>> University Professor
>>> Physics Department
>>> Old Dominion University
>>> Norfolk, VA 23529
>>> 757 683 5803
>>> 757 683 3038 (fax)
>>> weinstein at odu.edu <mailto:weinstein at odu.edu>
>>> http://www.lions.odu.edu/~lweinste/ <http://www.lions.odu.edu/%7Elweinste/>
>>
>> ------------------------------------------------------------------------
>>
>> Spam
>> <https://www.spamtrap.odu.edu/b.php?i=438400204&m=903a77671a30&t=20110126&c=s>
>> Not spam
>> <https://www.spamtrap.odu.edu/b.php?i=438400204&m=903a77671a30&t=20110126&c=n>
>> Forget previous vote
>> <https://www.spamtrap.odu.edu/b.php?i=438400204&m=903a77671a30&t=20110126&c=f>
>
> --
> Sincerely,
> Larry
>
> -----------------------------------------------------------
> Lawrence Weinstein
> University Professor
> Physics Department
> Old Dominion University
> Norfolk, VA 23529
> 757 683 5803
> 757 683 3038 (fax)
> weinstein at odu.edu <mailto:weinstein at odu.edu>
> http://www.lions.odu.edu/~lweinste/ <http://www.lions.odu.edu/%7Elweinste/>