[Dsg-mvt] MVT Mixing System Testing
Marc McMullen
mcmullen at jlab.org
Fri Dec 8 13:09:00 EST 2017
Hi Amrit,
On 12/6 the Saclay MVT gas controls started with demanding mixed gas
from the Hall B gas controls, which was setup in automatic mode. This
started the flow of Ar and C4H10 from the MVT mixing system. That
evening MVT staff observed fluctuations in flow from the C4H10 mass flow
controllers. We checked it out and noticed fluctuations in C4H10
supply, and odd behavior from the mfcs such as: reporting significant
flow, with the internal valve closed. extremely slow response to change
in flow commands. These issues have only been observed with the C4H10
mfcs. We spent most of the day yesterday troubleshooting the Mix 2 C4H10
mfc. The last part of the day we operated the mfc using the software
provided by MKS and performed the "zeroing" procedure a couple of times
to correct a flow offset. During this part of the procedure the gas
supply is mechanically valved off and the mfc is isolated as much as the
valve panel will allow. Generally, this produces a small flow value (0,
+/- < 1 unit), close to zero. As I observed the flow, I noted that the
flow was all over the place for a long time before it settled at a point
where I thought it was ready to complete the procedure. However once I
opened the manual gas valve the offset returned. This morning I noted
that it was back in the "normal" range, so that offset flow number must
take a while to bleed off.
In any case, we will start a rigorous testing procedure on the spare,
using a normal gas type (maybe compressed air) in a controlled
environment on Monday. Once we are sure that it works, we will have it
installed and perform those tests on the mixing system, first using
Argon and then finally C4H10. I should have a full report completed for
the next DSG status meeting on Wednesday.
We have informed the MVT group to use the 2 premix bottles they have on
the target gas pad for data taking. We have ordered 10 bottles of
premix, and it has been signed and ordered by procurement.
--
Marc McMullen
Detector Support Group
757-269-7738