[Dsg-mvt] MVT Mixing System Testing
Amrit Yegneswaran
yeg at jlab.org
Fri Dec 8 14:02:01 EST 2017
marc thanks for the detailed info.
as you say we'll discuss the matter in detail on wed.
check the power to the MFC .
i agree with the suggestion that the MVT group should proceed with the
pre-mixed system, while george, brian and you debug the mixing system.
amrit
On 12/8/2017 1:09 PM, Marc McMullen wrote:
> Hi Amrit,
>
> On 12/6 the Saclay MVT gas controls started with demanding mixed gas
> from the Hall B gas controls, which was setup in automatic mode. This
> started the flow of Ar and C4H10 from the MVT mixing system. That
> evening MVT staff observed fluctuations in flow from the C4H10 mass
> flow controllers. We checked it out and noticed fluctuations in C4H10
> supply, and odd behavior from the mfcs such as: reporting significant
> flow, with the internal valve closed. extremely slow response to
> change in flow commands. These issues have only been observed with
> the C4H10 mfcs. We spent most of the day yesterday troubleshooting the
> Mix 2 C4H10 mfc. The last part of the day we operated the mfc using
> the software provided by MKS and performed the "zeroing" procedure a
> couple of times to correct a flow offset. During this part of the
> procedure the gas supply is mechanically valved off and the mfc is
> isolated as much as the valve panel will allow. Generally, this
> produces a small flow value (0, +/- < 1 unit), close to zero. As I
> observed the flow, I noted that the flow was all over the place for a
> long time before it settled at a point where I thought it was ready to
> complete the procedure. However once I opened the manual gas valve the
> offset returned. This morning I noted that it was back in the "normal"
> range, so that offset flow number must take a while to bleed off.
>
> In any case, we will start a rigorous testing procedure on the spare,
> using a normal gas type (maybe compressed air) in a controlled
> environment on Monday. Once we are sure that it works, we will have it
> installed and perform those tests on the mixing system, first using
> Argon and then finally C4H10. I should have a full report completed
> for the next DSG status meeting on Wednesday.
>
> We have informed the MVT group to use the 2 premix bottles they have
> on the target gas pad for data taking. We have ordered 10 bottles of
> premix, and it has been signed and ordered by procurement.
>