_*Before all explanations, I want to emphasize that Maxence and I are
not gas experts. So we are just explaining you what we did,why we think
it seems to be safe to make another test of the mixing system with
isobutane. We decline any responsability if something goes wrong and
final approval to start using mixing system with isobutane must come
from the JLab staff in charge of it. *_
Last night with Maxence, we did the following procedure:
- Closed all the valves.
- Closed the manometer for the argon supply.
From this state where everything is closed, we opened the valve in the
rack on the isobutane line supply and then filled the line from the
isobutane bottle to the upstream valve of MFCs with isobutane. The
pressure was about 25psi (cannot be more precise since the heat tape was
ON at 15% at that moment... when heat tape is warming, pressure is
increasing in the line).
Then we closed the bottle and left the system in this state overnight,
i.e. all the line insulated with heat tape around at 15% filled with
isobutane.
This morning we went to purge the line. To do so, we first open the
relief valve and then we open the one of the valves allowing to purge
with argon the isobutane MFC line. It is important to first open the
relief valve, otherwise isobutane will first flow in a line which is not
insulated and condense within a few seconds. (you tried it yesterday
evening and relief valve was then spitting liquid and gaz. So 30F with
no insulation will make isobutane condensing quite fast.)
So this morning, only gaseous isobutane was exiting the relief valve.
Apparently it seems that isobutane is not condensing anymore. And once
the relief valve is closed, pressure in the line stayed at 0 (With
liquid in it, I would expect to have some pressure built in the line). I
nevertheless insist on the "apparently" since the relief valve is not
exactly on the isobutane line.
So, considering all this, I would _*only recommend*_ to start testing
the mixing system with isobutane again. But _*final approval must be
given either by a staff of JLab since this system is JLab responsability.*_
If you want a demonstration of the test later today (because the mail
might not be very clear without the rack in front of you), we can
organize it. However, I have not a faintest idea about what the next
test is to ensure that isobutane is gaseous, except going through one
MFC (no need to start using both immediately and risk to lose both of them).
Finally, 15% for the heat tape seems fine for ~30F. But temperature will
get colder over the week end and next week. We might consider to run
safe, i.e. testing again isobutane state as soon as we hit a new minimum
of temperature.
Kind regards,
Maxime