Maxime,
See if you can get the people together for a show and tell. I would be
interested in attending.
Volker
On 1/10/18 10:18 AM, mdefurne wrote:
>
> Dear all,
>
> _*Before all explanations, I want to emphasize that Maxence and I are
> not gas experts. So we are just explaining you what we did,why we
> think it seems to be safe to make another test of the mixing system
> with isobutane. We decline any responsability if something goes wrong
> and final approval to start using mixing system with isobutane must
> come from the JLab staff in charge of it. *_
>
> Last night with Maxence, we did the following procedure:
>
> - Closed all the valves.
>
> - Closed the manometer for the argon supply.
>
> From this state where everything is closed, we opened the valve in the
> rack on the isobutane line supply and then filled the line from the
> isobutane bottle to the upstream valve of MFCs with isobutane. The
> pressure was about 25psi (cannot be more precise since the heat tape
> was ON at 15% at that moment... when heat tape is warming, pressure is
> increasing in the line).
> Then we closed the bottle and left the system in this state overnight,
> i.e. all the line insulated with heat tape around at 15% filled with
> isobutane.
>
> This morning we went to purge the line. To do so, we first open the
> relief valve and then we open the one of the valves allowing to purge
> with argon the isobutane MFC line. It is important to first open the
> relief valve, otherwise isobutane will first flow in a line which is
> not insulated and condense within a few seconds. (you tried it
> yesterday evening and relief valve was then spitting liquid and gaz.
> So 30F with no insulation will make isobutane condensing quite fast.)
>
> So this morning, only gaseous isobutane was exiting the relief valve.
> Apparently it seems that isobutane is not condensing anymore. And once
> the relief valve is closed, pressure in the line stayed at 0 (With
> liquid in it, I would expect to have some pressure built in the line).
> I nevertheless insist on the "apparently" since the relief valve is
> not exactly on the isobutane line.
>
> So, considering all this, I would _*only recommend*_ to start
> testing the mixing system with isobutane again. But _*final approval
> must be given either by a staff of JLab since this system is JLab
> responsability.*_
>
> If you want a demonstration of the test later today (because the mail
> might not be very clear without the rack in front of you), we can
> organize it. However, I have not a faintest idea about what the next
> test is to ensure that isobutane is gaseous, except going through one
> MFC (no need to start using both immediately and risk to lose both of
> them).
>
> Finally, 15% for the heat tape seems fine for ~30F. But temperature
> will get colder over the week end and next week. We might consider to
> run safe, i.e. testing again isobutane state as soon as we hit a new
> minimum of temperature.
>
> Kind regards,
>
> Maxime
>