[Dsg-mvt] Mix 1 Setpoint
    mdefurne 
    mdefurne at jlab.org
       
    Thu Jan 11 16:26:54 EST 2018
    
    
  
Hi Brian,
Can you please indeed set the maximum flow rate at 1391. We will never 
reach such value.
Also, can you set the max pressure in tank 2 at 15 psi since the 
isobutane pressure is set at 25psi upstream of the MFC?
Thank you,
Maxime
Le 11/01/2018 à 13:58, Brian Eng a écrit :
> Also, the full scale (which determines minimum flow) can be adjusted if need be.
>
> For that I would need to know what the maximum flow rate you'd need out of the Ar MFC.
>
> It can be set from 1391 - 2780 sccm, the min flow is 1.5% of that value, so 41.7 sccm as currently configured.
>
> I don't know if the MFC retains that value when switching gases, so we'd need to test that if you decide we should change the value.
>
> ----- Original Message -----
>> From: "Brian Eng" <beng at jlab.org>
>> To: dsg-mvt at jlab.org
>> Sent: Thursday, January 11, 2018 1:49:09 PM
>> Subject: [Dsg-mvt] Mix 1 Setpoint
>> Hi Maxime,
>>
>> The flow rate you guys are requesting is currently to low for the Ar MFC, so it
>> isn't flowing anything.
>>
>> I would suggest raising the FMT (or FTT) setpoint to something higher.
> _______________________________________________
> Dsg-mvt mailing list
> Dsg-mvt at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-mvt
    
    
More information about the Dsg-mvt
mailing list