[Dsg-rich] Third cable and connector on RICH LV boards' service connector
Chris Cuevas
cuevas at jlab.org
Wed Aug 9 16:36:19 EDT 2017
Hi Tyler,
So your interlock system monitors signals that are relatively slow (temperature, humidity, pressures).
All LV and HV boards will reside in one mainframe, and these types of slow signals (including the relay transition time) should simply be
connected to the mainframe global interlock which will shut off LV and HV boards. There is no need for extra wiring to the service panel
to shut off each group of LV.
The HV and LV WILL be monitored (EPICS I presume) and the HV will be disabled in a much faster fashion by setting trip points to the
output current level for each channel. The trip point resolution is programmable and can be set very close to the nominal operating point for
two or three maPMT per HV channel. Trip points could(should) be assigned to each LV channel as well for fast response shutdown.
Thanks,
-Chris
~~~~~~~~~~~~~~~~~
----- Original Message -----
From: "Tyler Lemon" <tlemon at jlab.org>
To: "C. Cuevas" <cuevas at jlab.org>
Cc: "Marco Contalbrigo" <mcontalb at fe.infn.it>, "dsg-rich" <dsg-rich at jlab.org>, "Benjamin Raydo" <braydo at jlab.org>
Sent: Wednesday, August 9, 2017 4:10:46 PM
Subject: Re: [Dsg-rich] Third cable and connector on RICH LV boards' service connector
Hello Chris,
My reply is inline below in blue.
Hi,
If the 'hardware interlock system' produces an interlock relay closure, and this closure is the logical OR of several signals, then the closure should be connected to the CAEN mainframe interlock input(CPU). Is this the case?
For the LV boards, the hardware interlock system will open a relay to break the short circuits required at the service connector to enable a group's output. This is described in on page 7 of the attached manual for the A2518 boards.
We are not using the interlock input on the CAEN mainframe because when asked, Valery said HV should be turned off before LV. Using the interlock input on the mainframe will shutoff both HV and LV at the same time.
The interlock signals on the A2518 boards control two groups of 4 low voltage outputs. Each output channel has over-voltage and over-current trip points,
We are not monitoring any current or voltage output of HV or LV. The hardware interlock system will only disable HV/LV in the event that one of the hardware interlock system sensors (temperature, humidity, N2 flow, cooling airflow, cooling tank pressure) goes out of the determined limits.
and I do not see the need to interface to each group of LV interlocks if there is a global interlock signal that will latch a disable condition.
We have programed the hardware interlock system in a way that disables all LV groups at once and all HV channels at once. The hardware interlock system does not disable/enable each individual group.
Is it really necessary to sequence the shutoff of HV and LV?
If not, we can change the hardware interlock system code to use the CAEN mainframe interlock input.
Best regards,
Tyler