[Dsg-rich] RICH mirror reflectivity procedure and Mirror 3 samples 21 and 22 results
Valery Kubarovsky
vpk at jlab.org
Thu Jul 13 15:29:41 EDT 2017
Hello Tyler,
Reflectance for light with length more than 300 nm has to be near 85-90% (see the attached example). As I can understand from your spreadsheet the column A corresponds to the reference photodiode. It means that all three columns (B,F and G) have to be very close to each other.
However the Calibration B is different from F and G. It means probably that something happened with the light alignment. It is the reason that your are getting 100% reflectance.
Regards,
Valery
> From: "Tyler Lemon" <tlemon at jlab.org>
> To: "dsg-rich" <dsg-rich at jlab.org>
> Sent: Thursday, July 13, 2017 11:52:54 AM
> Subject: [Dsg-rich] RICH mirror reflectivity procedure and Mirror 3 samples 21
> and 22 results
> Hello,
> We have completed the reflectivity tests for Mirror 3, samples 21 and 22.
> Attached are the Excel files of the results.
> The following is the procedure we followed:
> 1.) Arrange photodiodes for calibration run.
> 2.) Run calibration. Calibration scans wavelengths from 300 nm- 650 nm at 10 nm
> steps.
> 3.) Arrange photodiodes and mirror sample 22 for test runs.
> 4.) Run test #1. Test also scans wavelengths from 300 nm- 650 nm at 10 nm steps.
> 5.) Repeat test scan (Test #2).
> 6.) Replace sample 22 with sample 21.
> 7.) Run test #1. Test scans wavelengths from 300 nm- 650 nm at 10 nm steps.
> 8.) Repeat test scan (Test #2).
> Best regards,
> Tyler
> _______________________________________________
> Dsg-rich mailing list
> Dsg-rich at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-rich
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20170713/b3b0d85e/attachment-0002.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Screen Shot 2017-07-13 at 15.21.55.png
Type: image/png
Size: 176251 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20170713/b3b0d85e/attachment-0002.png>
More information about the Dsg-rich
mailing list