[Dsg-rich] Reflectivity test station - alignment tests
Valery Kubarovsky
vpk at jlab.org
Thu Jul 20 10:23:49 EDT 2017
> From: "Valery Kubarovsky" <vpk at jlab.org>
> To: "mirazita" <Marco.Mirazita at lnf.infn.it>
> Cc: dsg-rich at jlab.org
> Sent: Thursday, July 20, 2017 10:17:08 AM
> Subject: Re: [Dsg-rich] Reflectivity test station - alignment tests
> Marco,
> I attached the deuterium source spectrum of the lamp that we are using for the
> reflectivity test.
> As you see the intensity of our lamp is order of magnitude HIGHER at low
> wavelength.
> Regards,
> Valery
>> From: "mirazita" <Marco.Mirazita at lnf.infn.it>
>> To: dsg-rich at jlab.org
>> Sent: Thursday, July 20, 2017 9:13:40 AM
>> Subject: Re: [Dsg-rich] Reflectivity test station - alignment tests
>> Hi Tyler,
>> thank you for the data.
>> I attached some analysis of the these data.
>> The phodiode readout variations are much bigger than yesterday, but at
>> the end I estimated a systematic uncertainty on the reflectivity of
>> about +/-2% above 400 nm.
>> This is acceptable.
>> At lower wavelengths, the variations are bigger, but that's because the
>> intensity of the lamp is order of magnitudes smaller.
>> It would be nice if you could redo the same series of measurements on
>> one of the carbon fiber samples.
>> Marco
>> Il 2017-07-19 22:56 Tyler Lemon ha scritto:
>> > Hello,
>> > Attached is an Excel sheet containing the results from today's
>> > alignment tests of the reflectivity tests station. The sheet contains
>> > the three data sets for separate calibration and reflectivity tests of
>> > the control mirror. After the three sets of results, there are
>> > statistics showing the mean photodiode currents, reflectivity, and
>> > percent error for a wavelength.
>> > The procedure for taking each measurements was:
>> > 1.) Align test station to take calibration run.
>> > 2.) Take calibration run.
>> > 3.) Re-align test station to measure mirror.
>> > 4.) Take mirror measurement run.
>> > Between measurements, we noted that the lamp power remained at 10.65V
>> > and 3.84A. We did not turn the lamp or picoampmeter off between
>> > measurements.
>> > The calculations at the end of the data show that there is a higher
>> > error of all measurements compared to yesterday's tests when the test
>> > station was not modified between measurements. Yesterday, the errors
>> > for calibrations were under 1%. Today, the errors for calibrations
>> > were about 6%. There are some exceptions at smaller wavelengths (300nm
>> > - 400nm) where the error for the calibration is greater.
>> > The calculations also show that the control mirror's reflectivity
>> > measurements had an error of about 6% for most wavelengths. Again,
>> > there are some exceptions for reflectivity measurements at smaller
>> > wavelengths (300nm - 420nm) where the error for the reflectivity was
>> > greater. One major exception was that the reflectivity measurement at
>> > 420 nm had an error of ~20%. This measurement is highlighted in orange
>> > in the table.
>> > In conclusion, today's results further support that the main source of
>> > error in the test station is in how repeatable the test station
>> > optics's alignment is between data runs.
>> > Best regards,
>> > Tyler
>> > _______________________________________________
>> > Dsg-rich mailing list
>> > Dsg-rich at jlab.org
>> > https://mailman.jlab.org/mailman/listinfo/dsg-rich
>> _______________________________________________
>> Dsg-rich mailing list
>> Dsg-rich at jlab.org
>> https://mailman.jlab.org/mailman/listinfo/dsg-rich
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20170720/c0bee6ca/attachment-0002.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Deuterium_source_spectrum.pdf
Type: application/pdf
Size: 226596 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20170720/c0bee6ca/attachment-0002.pdf>
More information about the Dsg-rich
mailing list