[Dsg-rich] Reflectivity test station - alignment tests
Tyler Lemon
tlemon at jlab.org
Thu Jul 20 10:26:08 EDT 2017
Hello Valery and Marco,
Both of your statements are correct, but it seems you are talking about two separate lamps.
For the reflectivity tests we have performed, we used the quartz-tungsten halogen (QTH) lamp since that lamp provides light in the visible spectrum. For the QTH lamp, it is as Marco said where the intensity at smaller wavelengths is lower.
The test station also has a deuterium lamp that we are not using since it provides light in the UV spectrum. The information Valery mentioned is for the deuterium lamp we are not using.
Best regards,
Tyler
From: "Valery Kubarovsky" <vpk at jlab.org>
To: "Mirazita Marco" <Marco.Mirazita at lnf.infn.it>
Cc: "dsg-rich" <dsg-rich at jlab.org>
Sent: Thursday, July 20, 2017 10:17:08 AM
Subject: Re: [Dsg-rich] Reflectivity test station - alignment tests
Marco,
I attached the deuterium source spectrum of the lamp that we are using for the reflectivity test.
As you see the intensity of our lamp is order of magnitude HIGHER at low wavelength.
Regards,
Valery
From: "mirazita" <Marco.Mirazita at lnf.infn.it>
To: dsg-rich at jlab.org
Sent: Thursday, July 20, 2017 9:13:40 AM
Subject: Re: [Dsg-rich] Reflectivity test station - alignment tests
BQ_BEGIN
Hi Tyler,
thank you for the data.
I attached some analysis of the these data.
The phodiode readout variations are much bigger than yesterday, but at
the end I estimated a systematic uncertainty on the reflectivity of
about +/-2% above 400 nm.
This is acceptable.
At lower wavelengths, the variations are bigger, but that's because the
intensity of the lamp is order of magnitudes smaller.
It would be nice if you could redo the same series of measurements on
one of the carbon fiber samples.
Marco
Il 2017-07-19 22:56 Tyler Lemon ha scritto:
> Hello,
>
> Attached is an Excel sheet containing the results from today's
> alignment tests of the reflectivity tests station. The sheet contains
> the three data sets for separate calibration and reflectivity tests of
> the control mirror. After the three sets of results, there are
> statistics showing the mean photodiode currents, reflectivity, and
> percent error for a wavelength.
>
> The procedure for taking each measurements was:
>
> 1.) Align test station to take calibration run.
>
> 2.) Take calibration run.
>
> 3.) Re-align test station to measure mirror.
>
> 4.) Take mirror measurement run.
>
> Between measurements, we noted that the lamp power remained at 10.65V
> and 3.84A. We did not turn the lamp or picoampmeter off between
> measurements.
>
> The calculations at the end of the data show that there is a higher
> error of all measurements compared to yesterday's tests when the test
> station was not modified between measurements. Yesterday, the errors
> for calibrations were under 1%. Today, the errors for calibrations
> were about 6%. There are some exceptions at smaller wavelengths (300nm
> - 400nm) where the error for the calibration is greater.
>
> The calculations also show that the control mirror's reflectivity
> measurements had an error of about 6% for most wavelengths. Again,
> there are some exceptions for reflectivity measurements at smaller
> wavelengths (300nm - 420nm) where the error for the reflectivity was
> greater. One major exception was that the reflectivity measurement at
> 420 nm had an error of ~20%. This measurement is highlighted in orange
> in the table.
>
> In conclusion, today's results further support that the main source of
> error in the test station is in how repeatable the test station
> optics's alignment is between data runs.
>
> Best regards,
>
> Tyler
>
>
> _______________________________________________
> Dsg-rich mailing list
> Dsg-rich at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-rich
_______________________________________________
Dsg-rich mailing list
Dsg-rich at jlab.org
https://mailman.jlab.org/mailman/listinfo/dsg-rich
BQ_END
_______________________________________________
Dsg-rich mailing list
Dsg-rich at jlab.org
https://mailman.jlab.org/mailman/listinfo/dsg-rich