[Dsg-rich] [EXTERNAL] RICH electronics
Matteo Turisini
matt at jlab.org
Mon Sep 23 17:03:41 EDT 2019
Hi Tyler,
the high end of our FPGAs is 85 celsius.
70 celsius was adopted as a precaution in the alarm system (now reduced to 60?)
Ben can for sure argue better than me.
Cheers to everybody,
Matteo
> Il giorno 23 set 2019, alle ore 22:48, Amrit Yegneswaran <yeg at jlab.org> ha scritto:
>
> tyler
> address fpga temp issue with ben.
> i believe 60c is right at the edge of timing issues, not sure.
> From: Dsg-rich <dsg-rich-bounces at jlab.org> on behalf of Tyler Lemon <tlemon at jlab.org>
> Sent: Monday, September 23, 2019 3:11 PM
> To: Marco Mirazita <Marco.Mirazita at lnf.infn.it>
> Cc: dsg-rich <dsg-rich at jlab.org>
> Subject: Re: [Dsg-rich] [EXTERNAL] RICH electronics
>
> Hi Marco,
>
> Looking back at EP temperatures during steady operations (I picked 3/10/2019 to 3/31/2019), Hardware Interlock RTDs were ~40 deg C and FPGA temperatures were ~60 deg C.
>
> Is there a temperature increase that would be considered acceptable? For example if the FPGA temperatures were ~70 deg C rather than ~60 deg C?
>
> The first test should be determining what flow rate is actually needed to maintain EP temperatures in that acceptable range with all RICH electronics on.
>
> -Tyler
>
> From: Marco Mirazita <Marco.Mirazita at lnf.infn.it>
> Sent: Monday, September 23, 2019 9:01 AM
> To: Tyler Lemon <tlemon at jlab.org>
> Cc: dsg-rich <dsg-rich at jlab.org>
> Subject: Re: [Dsg-rich] [EXTERNAL] RICH electronics
>
> Hi Tyler,
> sorry, I had in mind that the replacement of the PT was scheduled for
> last Wednesday.
>
> The tests we have in mind to optimize the cooling system is basically
> what you said. But we still have to make a plan of the specific
> measurements we need. We will do this plan in the next weeks.
> Any suggestion from you and the DSG is welcome.
>
> Best regards,
> Marco
>
>
> Il 2019-09-20 20:46 Tyler Lemon ha scritto:
> > Hi Marco,
> >
> > The PT replacement has not been completed. It is scheduled for next
> > Wednesday September 25.
> >
> > As for the visit to optimize RICH running conditions, I will not be
> > at work for at least part of the week of October 21 to October 25,
> > however there should be other people from DSG that would be able to
> > help.
> >
> > What tests are being planned for the visit? Would it essentially be
> > varying the settings of the cooling system (flow rate, buffer tank
> > pressure, compressor output pressure) and seeing how the temperature
> > within RICH is affected?
> >
> > Best regards,
> > Tyler
> >
> > -------------------------
> >
> > FROM: Marco Mirazita <Marco.Mirazita at lnf.infn.it>
> > SENT: Friday, September 20, 2019 12:21 PM
> > TO: Tyler Lemon <tlemon at jlab.org>
> > CC: dsg-rich <dsg-rich at jlab.org>; Valery Kubarovsky <vpk at jlab.org>;
> > marco contalbrigo <mcontalb at fe.infn.it>
> > SUBJECT: Re: [EXTERNAL] RICH electronics
> >
> > Hi Tyler,
> > do you know if Bob has completed the replacement of the FT?
> > If so, the compressor could be restarted on Monday morning, once we
> > are
> > sure everything is working properly I can turn on the RICH
> > electronics
> > and keep it on for the night and then make the tests on Tuesday
> > morning.
> >
> > As you know, we didn't have much time to optimize the working point
> > of
> > the cooling system after the RICH installation, we simply choose
> > running
> > conditions that ensure a safe operation of the detector.
> > But these running conditions might be problematic with two RICH
> > modules
> > and definitely not sustainable in case of failure of one of the two
> > compressors.
> > Therefore, also in preparation of the upcoming CLAS12 data taking,
> > I'm
> > planning a visit to JLab for Dario and Sandro to try to optimize the
> > running conditions, and to be effective this work must be coordinated
> >
> > with you and the DSG.
> > The proposed dates are from October 21 to October 31. Is this plan ok
> >
> > for you?
> > Thank you,
> > Marco
> >
> > Il 2019-09-18 14:12 Tyler Lemon ha scritto:
> > > Hi Marco,
> > >
> > > Bob Miller has scheduled the replacement of the faulty pressure
> > > transducer (PT) on the air-cooling buffer tank for next Wednesday,
> > > September 25.
> > >
> > > Could the tests be scheduled to start Thursday September 26?
> > >
> > > Best regards,
> > > Tyler
> > >
> > > -------------------------
> > >
> > > FROM: Marco Mirazita <Marco.Mirazita at lnf.infn.it>
> > > SENT: Wednesday, September 18, 2019 4:08 AM
> > > TO: Tyler Lemon <tlemon at jlab.org>
> > > CC: dsg-rich <dsg-rich at jlab.org>; Valery Kubarovsky <vpk at jlab.org>;
> > > marco contalbrigo <mcontalb at fe.infn.it>
> > > SUBJECT: [EXTERNAL] RICH electronics
> > >
> > > Hi Tyler,
> > > the RICH electronics is off since quite a lot of time and in
> > > preparation
> > > of the upcoming CLAS12 running I would like to turn it on to verify
> > > that
> > > everything is still working and eventually to take new scaler
> > > calibration data.
> > > I could plan this work for the beginning of the next week.
> > > In preparation for that, we should coordinate to have all the
> > systems
> > >
> > > ready. Would you be available to start the compressor for the test?
> > > Thank you,
> > > Marco
> >
> > _______________________________________________
> > Dsg-rich mailing list
> > Dsg-rich at jlab.org
> > https://mailman.jlab.org/mailman/listinfo/dsg-rich
> _______________________________________________
> Dsg-rich mailing list
> Dsg-rich at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-rich