[dsg-hallb_rich] HallB open for RICH work

Amrit Yegneswaran yeg at jlab.org
Thu Jul 17 21:35:49 EDT 2025


good work brian
why would the averaging stop
isnt it a software feature
its like saying i swapped my mac and now my software doesnt work
me thinks someone goofed with the software and did not log it
sgree?

Get Outlook for iOS<https://aka.ms/o0ukef>
________________________________
From: dsg-hallb_rich <dsg-hallb_rich-bounces at jlab.org> on behalf of Brian Eng via dsg-hallb_rich <dsg-hallb_rich at jlab.org>
Sent: Thursday, July 17, 2025 5:30:10 PM
To: Valery Kubarovsky <vpk at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>; Christopher Dilks <dilks at jlab.org>; Brian Eng <beng at jlab.org>
Subject: Re: [dsg-hallb_rich] HallB open for RICH work

So I started poking around the LabVIEW code running on the cRIO and I think I see the issue with the PVs being reported differently despite it seeming like the modules channel readout working correctly.

Some of the channels have averaging enabled with a sample size of 100. So for example on the N2 flow the raw signal is only moving around a few mV which when averaged over that sample size almost never changes (the handful of ones I've seen have been at the nV level). I suspect once this is converted to flow it doesn't result in any changes to the PV itself which is why it appears as if it's stuck.

I don't know why some channels/signals have averaging enabled and others don't, but maybe in one of the swaps/recoveries that was done something changed there?



There was a real issue with that particular cRIO (the disk errors and being unable to remotely reboot it), but I suspect a power cycle/hard reset would have recovered it.

________________________________
From: dsg-hallb_rich <dsg-hallb_rich-bounces at jlab.org> on behalf of Brian Eng via dsg-hallb_rich <dsg-hallb_rich at jlab.org>
Sent: Thursday, July 17, 2025 4:34 PM
To: Valery Kubarovsky <vpk at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>; Christopher Dilks <dilks at jlab.org>
Subject: Re: [dsg-hallb_rich] HallB open for RICH work

iocrichIntlk

I rebooted it, but as expected it didn't seem to change anything.

The the PV for N2 Volume PPM is constantly updating in EPICS while the N2 flow is "stuck" despite the module channel itself constantly moving around.
________________________________
From: Valery Kubarovsky <vpk at jlab.org>
Sent: Thursday, July 17, 2025 4:01 PM
To: Brian Eng <beng at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>; Christopher Dilks <dilks at jlab.org>
Subject: Re: HallB open for RICH work

What ioc are you talking about?
It is a good time, do it now.
________________________________
From: Brian Eng <beng at jlab.org>
Sent: Thursday, July 17, 2025 15:11
To: Valery Kubarovsky <vpk at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>; Christopher Dilks <dilks at jlab.org>
Subject: Re: HallB open for RICH work

Maybe we could reset the softIOC to see if that makes any difference?

I suspect it won't, but maybe it's worth a shot while the beam is still down?
________________________________
From: Valery Kubarovsky <vpk at jlab.org>
Sent: Thursday, July 17, 2025 3:06 PM
To: Brian Eng <beng at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>; Christopher Dilks <dilks at jlab.org>
Subject: Re: HallB open for RICH work


Hi Brian,


Thank you for taking the time to look into the issues with the cRIO. I really appreciate your effort to fix it even though there wasn’t enough time to complete the work. I will let you know when we will have the opportunity to continue troubleshooting.

Best regards,
Valery



 - Valery
________________________________
From: Brian Eng <beng at jlab.org>
Sent: Thursday, July 17, 2025 2:41:41 PM
To: Valery Kubarovsky <vpk at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>; Christopher Dilks <dilks at jlab.org>; Brian Eng <beng at jlab.org>
Subject: Re: HallB open for RICH work

I forgot to fill out the notify field so here's the logbook entry: https://logbooks.jlab.org/entry/4420154

________________________________
From: dsg-hallb_rich <dsg-hallb_rich-bounces at jlab.org> on behalf of Brian Eng via dsg-hallb_rich <dsg-hallb_rich at jlab.org>
Sent: Thursday, July 17, 2025 1:15 PM
To: Valery Kubarovsky <vpk at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>; Christopher Dilks <dilks at jlab.org>
Subject: Re: [dsg-hallb_rich] HallB open for RICH work

Hi All,

I was unable to get the spare fully online in time and so reverted back to the previous cRIO.

I'll post a log entry later.

There doesn't appear to be any planned maintenance next week so I'm not sure what the best course of action would be as the "simple" swap of controllers didn't work out. It will require additional troubleshooting which is extremely hard to estimate the length of time it will take so isn't really suitable to do during a short access window.

________________________________
From: dsg-hallb_rich <dsg-hallb_rich-bounces at jlab.org> on behalf of Christopher Dilks via dsg-hallb_rich <dsg-hallb_rich at jlab.org>
Sent: Thursday, July 17, 2025 8:50 AM
To: Valery Kubarovsky <vpk at jlab.org>; dsg-hallb_rich at jlab.org <dsg-hallb_rich at jlab.org>
Subject: [dsg-hallb_rich] Fw: HallB open for RICH work

Hi Valery and DSG,

See message below from Florian:
________________________________
From: Florian Hauenstein <hauenst at jlab.org>
Sent: Thursday, July 17, 2025 08:36
To: Christopher Dilks <dilks at jlab.org>; Brian Eng <beng at jlab.org>
Subject: HallB open for RICH work

Hi Brian,

HallB is open now for the RICH work.
We would like to have the work completed before noon so that we have enough access keys for a walkthrough in the afternoon.

Best,
Florian
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rich/attachments/20250718/26ee22a4/attachment-0001.htm>


More information about the dsg-hallb_rich mailing list