[Hallb-solenoid] Interlock Checkout List P027 Completed
Amrit Yegneswaran
yeg at jlab.org
Fri Oct 7 14:01:18 EDT 2016
quite exceptional!
compliments to both of you for having persevered.
From: "Pablo Campero" <campero at jlab.org>
To: "Ruben Fair" <rfair at jlab.org>
Cc: hallb-solenoid at jlab.org
Sent: Friday, October 7, 2016 1:44:21 PM
Subject: [Hallb-solenoid] Interlock Checkout List P027 Completed
Dear Ruben,
Tyler and I completed all tasks Hall B Pre-Power-Up Interlock Checkout Procedure- P027.
We noted failures in the following tasks:
1. "Verify interlock to prevent over current"
- EPICS would not allows any current set point to be entered, regardless of whether current limit is exceeded.
2. "Comparator 4 1st Threshold"
- Because of noise, interlock tripped continuously with a 1st threshold limit of 150 [mV].
- VT10-DAQ (used for comparator 4) had about 50 [mV] noise with voltage source connected, and 75 [mV] without voltage source connected.
3. "Comparator 4 2nd Threshold"
- Because of noise, interlock tripped when the difference between voltage sources was 28 [mV] (It should be about 200 [mV]).
4. "Comparator 6 1st Threshold"
- Because of noise, interlock tripped continuously with a 1st threshold limit of 150 [mV].
- VT6-DAQ (used for comparator 6) had about 50 [mV] noise with voltage source connected.
5. "Comparator 6 2nd Threshold"
- Because of noise, interlock tripped continuously with a 2nd threshold limit of 200 [mV].
6. "Test Danfysik QD Sum1", " Test Danfysik QD Sum2" and " Test Danfysik QD Sum 3"
- MPS can not be reset from EPICS.
- Had to reset from local MPS control panel .
Best Regards
Pablo Campero
_______________________________________________
Hallb-solenoid mailing list
Hallb-solenoid at jlab.org
https://mailman.jlab.org/mailman/listinfo/hallb-solenoid
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/hallb-solenoid/attachments/20161007/6b9698af/attachment-0001.html>
More information about the Hallb-solenoid
mailing list